  • in reply to: Disposable curtains/screens #70897
    Sue Finch
    Participant

    Author:
    Sue Finch

    Email:
    suzef@BIGPOND.NET.AU

    Organisation:

    State:

    Hi Maree

    I too have been around for many years in the Theatre and then into the CSSD and more recently in Infection Prevention and Control arena. Whilst I appreciate the rationale for the need for single use items and agree with the list you mentioned I too have a real concern for the ever increasing push to single use especially non critical items such as curtains and wonder where will it end? I am of the opinion there is a place for both single use and reusable products in our industry but we need to be more responsible in the choices we make and consider the impact these choices will make on future generations.

    Just a foot note these days the advances in technology mean water savings, recycle programs and add to that highly sophisticated machines with power saving attributes the environmental footprint generated in the reprocessing is minimal in the modern laundry.

    Clinical Nurse Consultant

    Somaia Group

    45 Holloway Drive Bayswater. Vic. 3153

    M: 0419306845

    E:sfinch@simbatex.com.au

    Hi Marie

    Great questions to ask. I would also like to point out that reusable curtains that require regular laundering or dry cleaning also have an environmental footprint, and this must be taken in consideration when looking at environmental impacts. Since this is often outside of the health facility, sometimes it is left out of the discussions. Laundries in particular are great users of energy and their environmental impact is massive.

    Cheers

    Michael;

    Michael Wishart

    Infection Control Coordinator

    Holy Spirit Northside Private Hospital

    627 Rode Road, Chermside, Qld 4032

    t: (07) 3326 3068 | f: (07) 3607 2226

    e: Michael.Wishart@hsn.org.au

    w: http://www.holyspiritnorthside.org.au

    Please consider the environment before printing this email

    I have been following the conversation in this thread and would like to raise an issue/question

    My organisation is constantly looking at ways to reduce its environmental footprint.

    We are a member of the Victorian Green Health Round Table Group, which provides a forum of information sharing for member organisations related to environmental concerns.

    One key activity undertaken by this group is comparing waste volumes generated with the goal to reduce.

    One of the major changes in providing health services since I began nursing (which was a good few years ago) is the increasing use of disposable items; from kimguard wraps, surgical drapes, suction tubing, endotracheal tubes, and surgical glovesthe list goes on.

    Some of these are absolutely no-brainers when it comes to rationale.

    Disposable curtains are large and take up a lot of space in a bin, hence in landfill, prior to breaking down.

    How do organisations who use disposable curtains weigh up this issue in contrast to the infection control risk related to privacy screens?

    Maree Sommerville

    Infection Control Coordinator

    Mercy Hospital for Women

    Heidelberg

    (03) 8458 4759

    _____

    Hi All,

    Have a small issue Disposable curtains/screens!

    Would appreciate feedback from areas that are using the disposable curtain/screens in their facilities

    The issue is around cost of linen vs disposable curtains/screens.

    We have trialed & like what we have but those who watch the pennies are questioning their use.

    Originally we brought them into our ED because the poor terminal cleaning staff were frantic with attending the cleaning ( which involves the replacement of curtains).

    The NUM of ED was indicating at this particular incident -that there were three ambulances waiting to off load patients onto ED beds which were being held up by the terminal cleaning required.

    Amongst other actions taken regarding this issue in ED-was the implementation of the disposable curtains.

    Now the question being asked is who else in other health areas has disposable curtains/screens & where are they ( ie high risk areas).

    Much appreciate any assistance with this.

    Thank you

    Vicki Denyer

    Clinical Nurse Consultant | Infection Prevention & Control Unit
    Lismore Base Hospital
    Tel 02 6620 2385 | vicki.denyer@ncahs.health.nsw.gov.au

    _____

    This message is intended for the addressee(s) named and may contain confidential information. If you are not the intended recipient, please delete the message and any attachments and notify the sender. Views expressed in this message are those of the individual sender, and are not necessarily the views of NSW Health or any of its entities.

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    DISCLAIMER This email and any files transmitted with it
    may be confidential and intended solely for the use of the
    individual or entity to whom they are addressed. If you have
    received this email in error please notify the sender immediately
    by return email and delete or destroy this message and its
    attachments. While this email and any attachments have
    been cleared by Mercy Health’s virus protection systems, recipients
    should use their own systems to detect computer viruses or other
    materials that may corrupt files or systems. Find out more about
    Mercy Health at http://www.mercyhealth.com.au

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    _____

    Message protected by MailGuard: e-mail anti-virus, anti-spam and content filtering.
    http://www.mailguard.com.au

    Report this message as spam

    WARNING : This email contains information, which is CONFIDENTIAL, and that maybe subject to LEGAL PRIVILEGE. This e-mail and any attachments to it (the “Communication”) is confidential and is for the use only of the intended recipient, and may not duplicated or used by any other party without the express consent of the sender. The Communication may contain copyright material of St Vincent’s Health & Aged Care(“SVHAC”), or any of its related entities or of third parties. If you are not the intended recipient of the Communication, please notify the sender immediately by return e-mail, delete the Communication, and do not read, copy, print, retransmit, store or act in reliance on the Communication. Any views expressed in the Communication are those of the individual sender only, unless expressly stated to be those of SVHAC. SVHAC does not guarantee the integrity of the Communication, or that it is free from errors, viruses or interference. Thank-you.

    _____

    Message protected by MailGuard: e-mail anti-virus, anti-spam and content filtering.
    http://www.mailguard.com.au

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    in reply to: ENDOSCOPY #70284
    Sue Finch
    Participant

    Author:
    Sue Finch

    Email:
    suzef@BIGPOND.NET.AU

    Organisation:

    State:

    Hi Jenny

    As Brand names are discouraged from being used in this forum I will be very
    generic so hope it is understandable.

    I suspect you have already had a look at this online instructional video for
    the mechanical device you refer to, but if not I have included the link for
    you and others to look at.
    http://legacy.minntech.com/medivators/resource/video/scope-buddy.htm.

    This is an American product and as such some of the language and practices
    in the video are not reflective of our National Endoscopy Guidelines, such
    as leaving the decontaminated tubing full of disinfectant attached to the
    machine to soak for extended periods or as an alternative rinsed tubing can
    be hung to dry.

    GENCA guidelines -Infection Control in Endoscopy Guidelines states
    -cleaning adaptors and flushing systems must be sterilised via steam or
    approved low temperature equivalent or at least thermally disinfected after
    each use. I would suggest leaving tubing soaking for an extended period of
    time would pose Infection Prevention risks particularly any of the
    resistive water borne microorganisms. If the manufacturer is suggesting the
    tubing be disinfected in accordance with routine disinfection instructions
    after each use then that would make more sense, however this would require
    enough tubing to allow cleaning, rinsing and disinfecting for each scope.

    The disinfectant that the manufacturer is using in the video is
    glutaraldehyde, so assume a similar glutaraldehyde chemical is being used in
    your department. Glutaraldehyde 2% is an approved TGA high level
    disinfectant but no where can I find the efficacy of items after prolonged
    soaking which suggests that this is not its intended use. Plenty of
    information however out there about the hazards associated with the use of
    glutaraldehyde which is a concern.

    I would suggest the sterilization of the tubing via low temperature or steam
    after each use or the cleaning and high level disinfection then rinsing in
    between use. I would be asking the manufacturer to supply instructions on
    how these reusable medical devices should be appropriately processed for
    reuse. If they cannot supply them to you then you can notify the TGA and
    report this as this is a mandatory requirement of manufacturers and also
    Sponsors of Overseas manufactures (suppliers) of Reusable Medical
    Devices(RMD). But in the interim obviously disinfection via soaking is the
    only option recommended by the manufacturer so according to GENCA
    guidelines it must e performed in between each use and must be subject to
    the same requirements of documentation as required for the scope. For
    terminal processing I would suggest the tubing be thoroughly dried in a
    drying cupboard (not air dried)before storing in a manner that prevents
    contamination.

    Regards

    Sue Finch

    Nurse Consultant

    Infection Prevention & Control

    Sterilizing Services

    P: 0419306845

    E: suzef@bigpond.net.au

    Of Jenny McCarthy

    Dear all

    We have recently put a “scope buddy” in our endoscopy area (a device that
    automatically flushes the scope prior to putting in the AFER) . My concern
    is the suppliers have told the endoscopy staff that “other hospitals” leave
    the tubing attached to the scope buddy with the low level instrument grade
    disinfectant through the tubing.

    The tubing is reusable and staff have set up a system where the tubing sits
    in the disinfectant until the next list. Any comments?

    Thanks
    Jenny
    Jenny McCarthy
    OR Manager/Infection Prevention and Control Coordinator
    Maryvale Private Hospital

    Maryvale Private Hospital Confidentiality and Privacy Notice

    NOTICE – This e-mail is only intended to be read by the named recipient. It
    may contain information which is confidential or proprietary. If you are not
    the intended recipient please notify the sender immediately and delete this
    e-mail. If you are not the intended recipient you should not copy this
    e-mail or use the information contained in it for any purpose nor disclose
    its contents to any other person. Maryvale Private Hospital Pty Ltd accepts
    no responsibility for electronic viruses or damage caused as a result of
    this email or for changes made to this email or to any attachments after
    transmission from SIAG. You should not distribute or publish the contents of
    this email or any attachment without the prior consent of Maryvale Private
    Hospital.
    P Please consider the environment before printing this email

    Messages posted to this list are solely the opinion of the authors, and do
    not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives
    registration and login required.

    Replies to this message will be directed back to the list. To create a new
    message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to
    aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without
    the quotes) to listserv@aicalist.org.au

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    in reply to: Sterile instrument storage systems #70096
    Sue Finch
    Participant

    Author:
    Sue Finch

    Email:
    suzef@BIGPOND.NET.AU

    Organisation:

    State:

    Hi Jennifer

    Yes the standards do require that shelving is 250mm from the floor and 440mm
    from ceiling and light fittings to allow free circulation of air, prevent
    heating from light fittings and of course allows the cleaning of the store
    room floor as well. The draft of the new AS 4187 standard just emphasises
    the critical points of being far enough away from the ceiling, walls and
    floor to allow cleaning without compromising stock, allow adequate
    circulation of air and stock is prevented from being crushed, bent or
    punctured etc As for the materials for the storage shelves, AS 4187-2003
    suggests the ideal materials are stainless steel, chromium plated or plastic
    coated mesh because they avoid the collection of dust and are easily
    cleaned, but smooth laminated surfaces are also acceptable providing that
    they are kept free from dust and vermin. The important facts for storage
    surfaces is that they are easily cleaned, maintains the stock in a cool and
    dry condition, does not compromise stock sterility and is used for the
    dedicated purpose of sterile stock storage.
    Hope this helps

    Sue Finch
    Nurse Consultant Infection Prevention

    —–Original Message—–
    Of Jennifer Cartwright

    Dear all,

    I am currently researching current standards for storing sterile reusable
    medical devices RMD’s. AS4187 standards require RMD’s to be stored 250mm off
    the floor, require good ventilation, easily cleaned smooth surfaces,
    temperature control and a dedicated area. The containers the instruments are
    placed in are also to be easily cleaned and no rough edges etc. I was of the
    opinion that the storage containers should be open wire baskets, however I
    can’t find a source for this and wanted to see if anyone knew where I can
    find the standards required by the health department? Furthermore should the
    shelving be open wire shelving as well. I have not been able to find any
    source that specifically states what type of system is required.

    Kind regards

    Jennifer Cartwright
    Infection Control
    Eye Surgery Foundation
    Perth WA

    Sent from my iPad
    Messages posted to this list are solely the opinion of the authors, and do
    not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives
    registration and login required.

    Replies to this message will be directed back to the list. To create a new
    message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to
    aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without
    the quotes) to listserv@aicalist.org.au

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

Viewing 3 posts - 1 through 3 (of 3 total)