Hi Maree
I too have been around for many years in the Theatre and then into the CSSD and more recently in Infection Prevention and Control arena. Whilst I appreciate the rationale for the need for single use items and agree with the list you mentioned I too have a real concern for the ever increasing push to single use especially non critical items such as curtains and wonder where will it end? I am of the opinion there is a place for both single use and reusable products in our industry but we need to be more responsible in the choices we make and consider the impact these choices will make on future generations.
Just a foot note these days the advances in technology mean water savings, recycle programs and add to that highly sophisticated machines with power saving attributes the environmental footprint generated in the reprocessing is minimal in the modern laundry.
Clinical Nurse Consultant
Somaia Group
45 Holloway Drive Bayswater. Vic. 3153
M: 0419306845
E:sfinch@simbatex.com.au
Hi Marie
Great questions to ask. I would also like to point out that reusable curtains that require regular laundering or dry cleaning also have an environmental footprint, and this must be taken in consideration when looking at environmental impacts. Since this is often outside of the health facility, sometimes it is left out of the discussions. Laundries in particular are great users of energy and their environmental impact is massive.
Cheers
Michael;
Michael Wishart
Infection Control Coordinator
Holy Spirit Northside Private Hospital
627 Rode Road, Chermside, Qld 4032
t: (07) 3326 3068 | f: (07) 3607 2226
w: http://www.holyspiritnorthside.org.au
Please consider the environment before printing this email
I have been following the conversation in this thread and would like to raise an issue/question
My organisation is constantly looking at ways to reduce its environmental footprint.
We are a member of the Victorian Green Health Round Table Group, which provides a forum of information sharing for member organisations related to environmental concerns.
One key activity undertaken by this group is comparing waste volumes generated with the goal to reduce.
One of the major changes in providing health services since I began nursing (which was a good few years ago) is the increasing use of disposable items; from kimguard wraps, surgical drapes, suction tubing, endotracheal tubes, and surgical glovesthe list goes on.
Some of these are absolutely no-brainers when it comes to rationale.
Disposable curtains are large and take up a lot of space in a bin, hence in landfill, prior to breaking down.
How do organisations who use disposable curtains weigh up this issue in contrast to the infection control risk related to privacy screens?
Maree Sommerville
Infection Control Coordinator
Mercy Hospital for Women
Heidelberg
(03) 8458 4759
_____
Hi All,
Have a small issue Disposable curtains/screens!
Would appreciate feedback from areas that are using the disposable curtain/screens in their facilities
The issue is around cost of linen vs disposable curtains/screens.
We have trialed & like what we have but those who watch the pennies are questioning their use.
Originally we brought them into our ED because the poor terminal cleaning staff were frantic with attending the cleaning ( which involves the replacement of curtains).
The NUM of ED was indicating at this particular incident -that there were three ambulances waiting to off load patients onto ED beds which were being held up by the terminal cleaning required.
Amongst other actions taken regarding this issue in ED-was the implementation of the disposable curtains.
Now the question being asked is who else in other health areas has disposable curtains/screens & where are they ( ie high risk areas).
Much appreciate any assistance with this.
Thank you
Vicki Denyer
Clinical Nurse Consultant | Infection Prevention & Control Unit
Lismore Base Hospital
Tel 02 6620 2385 | vicki.denyer@ncahs.health.nsw.gov.au_____
This message is intended for the addressee(s) named and may contain confidential information. If you are not the intended recipient, please delete the message and any attachments and notify the sender. Views expressed in this message are those of the individual sender, and are not necessarily the views of NSW Health or any of its entities.
MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.
The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.
Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.
Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au
To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.
You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au
DISCLAIMER This email and any files transmitted with it
may be confidential and intended solely for the use of the
individual or entity to whom they are addressed. If you have
received this email in error please notify the sender immediately
by return email and delete or destroy this message and its
attachments. While this email and any attachments have
been cleared by Mercy Health’s virus protection systems, recipients
should use their own systems to detect computer viruses or other
materials that may corrupt files or systems. Find out more about
Mercy Health at http://www.mercyhealth.com.auMESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.
The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.
Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.
Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au
To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.
You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au
_____
Message protected by MailGuard: e-mail anti-virus, anti-spam and content filtering.
http://www.mailguard.com.auReport this message as spam
WARNING : This email contains information, which is CONFIDENTIAL, and that maybe subject to LEGAL PRIVILEGE. This e-mail and any attachments to it (the “Communication”) is confidential and is for the use only of the intended recipient, and may not duplicated or used by any other party without the express consent of the sender. The Communication may contain copyright material of St Vincent’s Health & Aged Care(“SVHAC”), or any of its related entities or of third parties. If you are not the intended recipient of the Communication, please notify the sender immediately by return e-mail, delete the Communication, and do not read, copy, print, retransmit, store or act in reliance on the Communication. Any views expressed in the Communication are those of the individual sender only, unless expressly stated to be those of SVHAC. SVHAC does not guarantee the integrity of the Communication, or that it is free from errors, viruses or interference. Thank-you.
_____
Message protected by MailGuard: e-mail anti-virus, anti-spam and content filtering.
http://www.mailguard.com.auMESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.
The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.
Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.
Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au
To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.
You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au
MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.
The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.
Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.
Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au
To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.
You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au
Hi Jenny
As Brand names are discouraged from being used in this forum I will be very
generic so hope it is understandable.I suspect you have already had a look at this online instructional video for
the mechanical device you refer to, but if not I have included the link for
you and others to look at.
http://legacy.minntech.com/medivators/resource/video/scope-buddy.htm.This is an American product and as such some of the language and practices
in the video are not reflective of our National Endoscopy Guidelines, such
as leaving the decontaminated tubing full of disinfectant attached to the
machine to soak for extended periods or as an alternative rinsed tubing can
be hung to dry.GENCA guidelines -Infection Control in Endoscopy Guidelines states
-cleaning adaptors and flushing systems must be sterilised via steam or
approved low temperature equivalent or at least thermally disinfected after
each use. I would suggest leaving tubing soaking for an extended period of
time would pose Infection Prevention risks particularly any of the
resistive water borne microorganisms. If the manufacturer is suggesting the
tubing be disinfected in accordance with routine disinfection instructions
after each use then that would make more sense, however this would require
enough tubing to allow cleaning, rinsing and disinfecting for each scope.The disinfectant that the manufacturer is using in the video is
glutaraldehyde, so assume a similar glutaraldehyde chemical is being used in
your department. Glutaraldehyde 2% is an approved TGA high level
disinfectant but no where can I find the efficacy of items after prolonged
soaking which suggests that this is not its intended use. Plenty of
information however out there about the hazards associated with the use of
glutaraldehyde which is a concern.I would suggest the sterilization of the tubing via low temperature or steam
after each use or the cleaning and high level disinfection then rinsing in
between use. I would be asking the manufacturer to supply instructions on
how these reusable medical devices should be appropriately processed for
reuse. If they cannot supply them to you then you can notify the TGA and
report this as this is a mandatory requirement of manufacturers and also
Sponsors of Overseas manufactures (suppliers) of Reusable Medical
Devices(RMD). But in the interim obviously disinfection via soaking is the
only option recommended by the manufacturer so according to GENCA
guidelines it must e performed in between each use and must be subject to
the same requirements of documentation as required for the scope. For
terminal processing I would suggest the tubing be thoroughly dried in a
drying cupboard (not air dried)before storing in a manner that prevents
contamination.Regards
Sue Finch
Nurse Consultant
Infection Prevention & Control
Sterilizing Services
P: 0419306845
Of Jenny McCarthy
Dear all
We have recently put a “scope buddy” in our endoscopy area (a device that
automatically flushes the scope prior to putting in the AFER) . My concern
is the suppliers have told the endoscopy staff that “other hospitals” leave
the tubing attached to the scope buddy with the low level instrument grade
disinfectant through the tubing.The tubing is reusable and staff have set up a system where the tubing sits
in the disinfectant until the next list. Any comments?Thanks
Jenny
Jenny McCarthy
OR Manager/Infection Prevention and Control Coordinator
Maryvale Private HospitalMaryvale Private Hospital Confidentiality and Privacy Notice
NOTICE – This e-mail is only intended to be read by the named recipient. It
may contain information which is confidential or proprietary. If you are not
the intended recipient please notify the sender immediately and delete this
e-mail. If you are not the intended recipient you should not copy this
e-mail or use the information contained in it for any purpose nor disclose
its contents to any other person. Maryvale Private Hospital Pty Ltd accepts
no responsibility for electronic viruses or damage caused as a result of
this email or for changes made to this email or to any attachments after
transmission from SIAG. You should not distribute or publish the contents of
this email or any attachment without the prior consent of Maryvale Private
Hospital.
P Please consider the environment before printing this emailMessages posted to this list are solely the opinion of the authors, and do
not represent the opinion of ACIPC.Archive of all messages are available at http://aicalist.org.au/archives –
registration and login required.Replies to this message will be directed back to the list. To create a new
message send an email to aicalist@aicalist.org.auTo send a message to the list administrator send an email to
aicalist-request@aicalist.org.au.You can unsubscribe from this list be sending ‘signoff aicalist’ (without
the quotes) to listserv@aicalist.org.auMessages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.
Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.
Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au
To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.
You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au
Hi Jennifer
Yes the standards do require that shelving is 250mm from the floor and 440mm
from ceiling and light fittings to allow free circulation of air, prevent
heating from light fittings and of course allows the cleaning of the store
room floor as well. The draft of the new AS 4187 standard just emphasises
the critical points of being far enough away from the ceiling, walls and
floor to allow cleaning without compromising stock, allow adequate
circulation of air and stock is prevented from being crushed, bent or
punctured etc As for the materials for the storage shelves, AS 4187-2003
suggests the ideal materials are stainless steel, chromium plated or plastic
coated mesh because they avoid the collection of dust and are easily
cleaned, but smooth laminated surfaces are also acceptable providing that
they are kept free from dust and vermin. The important facts for storage
surfaces is that they are easily cleaned, maintains the stock in a cool and
dry condition, does not compromise stock sterility and is used for the
dedicated purpose of sterile stock storage.
Hope this helpsSue Finch
Nurse Consultant Infection Prevention—–Original Message—–
Of Jennifer CartwrightDear all,
I am currently researching current standards for storing sterile reusable
medical devices RMD’s. AS4187 standards require RMD’s to be stored 250mm off
the floor, require good ventilation, easily cleaned smooth surfaces,
temperature control and a dedicated area. The containers the instruments are
placed in are also to be easily cleaned and no rough edges etc. I was of the
opinion that the storage containers should be open wire baskets, however I
can’t find a source for this and wanted to see if anyone knew where I can
find the standards required by the health department? Furthermore should the
shelving be open wire shelving as well. I have not been able to find any
source that specifically states what type of system is required.Kind regards
Jennifer Cartwright
Infection Control
Eye Surgery Foundation
Perth WASent from my iPad
Messages posted to this list are solely the opinion of the authors, and do
not represent the opinion of ACIPC.Archive of all messages are available at http://aicalist.org.au/archives –
registration and login required.Replies to this message will be directed back to the list. To create a new
message send an email to aicalist@aicalist.org.auTo send a message to the list administrator send an email to
aicalist-request@aicalist.org.au.You can unsubscribe from this list be sending ‘signoff aicalist’ (without
the quotes) to listserv@aicalist.org.auMessages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.
Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.
Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au
To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.
You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au
-
AuthorPosts