Select Page

Debbie Macqueen

Forum Replies Created

Viewing 1 post (of 1 total)
  • Author
    Posts
  • in reply to: Re bronchoscopes and ureteroscopes #68276
    Debbie Macqueen
    Participant

    Author:
    Debbie Macqueen

    Position:

    Organisation:

    State:

    Hello Wendy
    We have both intubating bronchoscopes and video bronchoscopes. Our
    intubating Bronchoscopes are Ethylene Oxide gas sterilised off site
    (wrapped), our video bronchoscopes are also wrapped but go through our
    Sterrad (wrapped). We also follow the GENCA micro testing guidelines. We
    have decided to follow the first guideline where the Bronchoscopes
    (including the Intubating Bronchs), Duodenoscopes and Cystoscopes are
    tested monthly – even though they are wrapped and sterilised. Some
    institutions follow the guideline that recommends 3 monthly testing of
    wrapped, sterilised scopes.
    My philosophy is to sterilise scopes that can tolerate a sterilising
    method. the companies are happy to assist your queries.
    I guess we err on the side of caution, so it will be interesting once
    the new Infection Control guidelines are released. We will probably need
    to alter our practice to conform with the new guidelines.
    Regards
    Debbie McQueen
    Endoscopy Coordinator
    Hobart Private Hospital
    phone : 03 62143158
    mobile: 0419320211
    email: debbie.macqueen@healthscope.com.au

    PPlease consider the environment before printing this message

    >>> “Beckingham, Wendy” 4/05/2010 8:32 am
    >>>
    Currently we have bronchoscopes that are cleaned and high level
    disinfected through the soluscope and micro tested as per the GENCA
    guidelines.

    We have another facility on our campus that uses the Sterrad to
    sterilize the bronchoscopes and ureteroscopes. We also do micro
    testing
    of both of these scopes.

    We are wondering what other facilities around Australia are currently
    doing in relation to instruments going through the STERRAD and micro
    testing.

    Wendy Beckingham
    CNC Infection Control
    The Canberra Hospital
    pager 50390 or phone 43695

    ———————————————————————–
    This email, and any attachments, may be confidential and also
    privileged. If you are not the intended recipient, please notify the
    sender and delete all copies of this transmission along with any
    attachments immediately. You should not copy or use it for any purpose,
    nor disclose its contents to any other person.
    ———————————————————————–

    Messages posted to this list are solely the opinion of the authors, and
    do not represent the opinion of AICA.
    Archive of all messages are available at
    http://aicalist.org.au/archives – registration and login required.
    Replies to this message will be directed back to the list. To create a
    new message send an email to aicalist@aicalist.org.au
    To send a message to the list administrator send an email to
    aicalist-request@aicalist.org.au.
    You can unsubscribe from this list be sending ‘signoff aicalist’
    (without the quotes) to listserv@aicalist.org.au

    IMPORTANT – This email contains confidential information intended only
    for the person named above
    and may be subject to legal privilege. If you are not the intended
    recipient, any disclosure, copying or use
    of this information is prohibited. Healthscope provides no guarantee
    that this communication is free of
    virus or that it has not been intercepted or interfered with. If you
    have received this email in error or have
    any other concerns regarding its transmission, please notify
    Postmaster@healthscope.com.au. You must
    destroy the original transmission and its contents. Any views expressed
    within this communication are
    those of the individual sender, except where the sender specifically
    states them to be the views of
    Healthscope. If this document is not required for record keeping
    purposes please consider the
    environment before storing or printing. This communication should not be
    copied or disseminated without
    permission.

    Please consider the environment before printing this message

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of AICA.
    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.
    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au
    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.
    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

Viewing 1 post (of 1 total)