Select Page

ResMed VPAP machines

Home Forums Infexion Connexion ResMed VPAP machines

 | Click to Receive Email Notifications of Posts
Viewing 1 post (of 1 total)
  • Author
    Posts
  • #77863
    Anonymous
    Inactive

    Author:
    Anonymous

    Organisation:

    State:

    Evening all

    Does anyone have an experience (or evidence to either support or deny) the
    use of ResMed VPAP machines during COVID? I understand in general NIV use
    is limited though I also know some centres internationally (due to
    equipment shortages and need to increase support) have modified these
    machines for use.

    This is the machine I am specifically seeking information for:
    https://support.resmed.com/en-au/ventilators/lumis-150-vpap-st/?_ga=2.55465771.185471130.1617169065-2125133923.1613687576

    Thanks
    Abby

    Abby Peacock-Smith
    Infection Prevention & Control
    WHO, Papua New Guinea
    abby.peacocksmith@gmail.com

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au

    To send a message to the list administrator send an email to admin@acipc.org.au

    You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au

Viewing 1 post (of 1 total)
  • The forum ‘Infexion Connexion’ is closed to new topics and replies.