Hello Matthew
AS4187 covers the issue of validation in detail.
For a high pre-vac steriliser the most basic validation, other than the process, would be the pressure and temperature – making sure you have saturated steam at the right temperature for the correct duration.
Table 7.2 covers the steam quality etc.
Hope this helps.
Cheers
Sent from LR Consulting Engineers
> On 18 Mar 2020, at 5:45 pm, Matthew Mason wrote:
>