Select Page

Re: Ultrasound transducers covers

Home Forums Infexion Connexion Re: Ultrasound transducers covers

 | Click to Receive Email Notifications of Posts
Viewing 2 posts - 1 through 2 (of 2 total)
  • Author
    Posts
  • #73848
    Glenys Harrington
    Participant

    Author:
    Glenys Harrington

    Position:
    Consultant

    Organisation:
    Infection Control Consultancy (ICC)

    State:

    Hi All,

    Adding to the discussion on transducer covers the current guidelines should
    be less open ended and read “use a single-use high-quality transducer cover
    that has been validated for use by the manufacturer of the transducer/s”.

    It is important to ensure that bacterial efficacy, fit and barrier integrity
    information can be provided by the manufacturer of such covers.

    If condoms are used the manufacturer (condoms) should also provide
    information confirming that the condoms are suitable for use on medical
    devices such as transducers.

    This will ensure your warranty is not voided and that there is no off-label
    use.

    Regards

    Glenys

    Glenys Harrington

    Infection Control Consultancy (ICC)

    P.O. Box 6385

    Melbourne

    Australia, 3004

    M: +61 404816434

    E: infexion@ozemail.com.au

    Of Cathryn Murphy

    Hi Cate

    This is an interesting issue. Whilst the guidelines say exactly what you
    have reproduced below I think the emphasis should be on single-use
    high-quality transducer cover rather than using a condom. My reasoning is
    based on two things. Firstly, to minimise risk I think it is best to use
    equipment specifically manufactured for a specific purpose ie. use a
    transducer cover that is purpose manufactured vs a condom which is
    manufactured for contraceptive purposes not infection control although there
    are obvious IC benefits and one could argue either side.

    Secondly, the issue of U/S reprocessing was on the agenda at the recent APIC
    meeting in Portland and it seems that US ICPs are moving much more towards
    improved higher-levelled reprocessing than was previously performed. This
    has come about by greater appreciation for the potential risk of
    transmission of conditions such as HPV and also the availability of
    relatively inexpensive quality systems of reprocessing that are suitable for
    US probes.

    As a fall-back position perhaps the manufacturer of the US transducer could
    recommend a cover that they know through testing works well with their
    probe. It’s an interesting position that ASUM has adopted regarding condom
    use and I wonder what evidence they used to support it.

    Hope this helps

    Cath

    Cathryn Murphy RN B. Photog MPH PhD CIC

    Chief Executive Officer & Creative Director

    Infection Control Plus Pty Ltd

    QLD, Australia

    E:
    Cath@infectioncontrolplus.com.au

    M: +61 428 154154

    W:
    http://www.infectioncontrolplus.com.au

    mailto:AICALIST@AICALIST.ORG.AU] On Behalf Of Cate Coffey

    Hi everyone,

    Seeking advice on ultrasound transducer covers. The Guidelines for
    Reprocessing Ultrasound Transducers regarding transducer/probe covers state
    All intracavity transducers should be covered with a single-use high-quality
    transducer cover. This may include some , but not all brands of condom,
    specific transducer covers or surgical drapes.

    Can you tell me how I determine which brands of condoms are suitable? Is
    there a criteria to assist?

    thanks

    Cate Coffey | Clinical Nurse Consultant

    Infection Prevention and Control Unit | Central Australia Health Service

    Northern Territory Government

    Alice Springs Hopsital, Gap Rd, Alice Springs

    GPO Box 2234, Suburb, NT Postcode

    p … 08 89517737

    e … cate.coffey@nt.gov.au
    http://www.nt.gov.au/health

    Our Vision: Better health outcomes for all Central Australians

    Our Values: Community at the Centre | Equity and Integrity | We are
    Accountable | We are Relevant Today and Ready for Tomorrow | We are
    Committed to High Quality Care | We Value our Partnerships

    Central Australia Health Service is a Smoke Free Workplace

    The information in this e-mail is intended solely for the addressee named.
    It may contain legally privileged or confidential information that is
    subject to copyright. If you are not the intended recipient you must not
    use, disclose copy or distribute this communication. If you have received
    this message in error, please delete the e-mail and notify the sender. No
    representation is made that this e-mail is free of viruses. Virus scanning
    is recommended and is the responsibility of the recipient.

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO
    NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is
    discouraged by ACIPC. If you wish to discuss specific reference to products
    or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at
    http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new
    message send an email to
    aicalist@aicalist.org.au

    To send a message to the list administrator send an email to
    aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without
    the quotes) to listserv@aicalist.org.au

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO
    NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is
    discouraged by ACIPC. If you wish to discuss specific reference to products
    or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives
    registration and login required.

    Replies to this message will be directed back to the list. To create a new
    message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to
    aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without
    the quotes) to listserv@aicalist.org.au

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    #73859
    Silvana Harding
    Participant

    Author:
    Silvana Harding

    Position:
    Clinical Manager

    Organisation:
    Ansell Healthcare

    State:
    VIC

    Hi all,

    I can confirm that Ansell condoms do not have an indication for transducer use and so cannot can supply a resource to support this use. We can however, provide a static condom viral penetration test consisting of a bacteriophage (A type of virus which infects bacteria as its host. It is also often refer to as phage). This can be supplied upon request.

    Any condom use for the aforementioned practice would need to be based on own policy or clinical best judgement.

    Kind regards,

    Silvana

    Silvana Harding RN
    Clinical Consultant ANZ Medical

    Ansell
    Level 3/678 Victoria Street
    Richmond VIC 3121
    Australia

    http://www.ansell.com

    [cid:image001.png@01D2FBB0.7F592450]

    [cid:image001.png@01D1D206.53417310] [Picture1]

    [join][cid:image004.png@01D1D206.53417310][cid:image005.png@01D1D206.53417310][cid:image006.png@01D1D206.53417310][cid:image007.png@01D1D206.53417310]https://www.youtube.com/watch?v=[cid:image009.jpg@01D1D206.53417310][cid:image012.jpg@01D2FBB1.10628180]

    Hi All,

    Adding to the discussion on transducer covers the current guidelines should be less open ended and read “use a single-use high-quality transducer cover that has been validated for use by the manufacturer of the transducer/s”.

    It is important to ensure that bacterial efficacy, fit and barrier integrity information can be provided by the manufacturer of such covers.

    If condoms are used the manufacturer (condoms) should also provide information confirming that the condoms are suitable for use on medical devices such as transducers.

    This will ensure your warranty is not voided and that there is no off-label use.

    Regards

    Glenys

    Glenys Harrington
    Infection Control Consultancy (ICC)
    P.O. Box 6385
    Melbourne
    Australia, 3004
    M: +61 404816434
    E: infexion@ozemail.com.au

    Hi Cate

    This is an interesting issue. Whilst the guidelines say exactly what you have reproduced below I think the emphasis should be on single-use high-quality transducer cover rather than using a condom. My reasoning is based on two things. Firstly, to minimise risk I think it is best to use equipment specifically manufactured for a specific purpose ie. use a transducer cover that is purpose manufactured vs a condom which is manufactured for contraceptive purposes not infection control although there are obvious IC benefits and one could argue either side.

    Secondly, the issue of U/S reprocessing was on the agenda at the recent APIC meeting in Portland and it seems that US ICPs are moving much more towards improved higher-levelled reprocessing than was previously performed. This has come about by greater appreciation for the potential risk of transmission of conditions such as HPV and also the availability of relatively inexpensive quality systems of reprocessing that are suitable for US probes.

    As a fall-back position perhaps the manufacturer of the US transducer could recommend a cover that they know through testing works well with their probe. It’s an interesting position that ASUM has adopted regarding condom use and I wonder what evidence they used to support it.

    Hope this helps
    Cath

    Cathryn Murphy RN B. Photog MPH PhD CIC
    Chief Executive Officer & Creative Director
    Infection Control Plus Pty Ltd
    QLD, Australia

    E: Cath@infectioncontrolplus.com.au
    M: +61 428 154154
    W:http://www.infectioncontrolplus.com.au

    Hi everyone,
    Seeking advice on ultrasound transducer covers. The Guidelines for Reprocessing Ultrasound Transducers regarding transducer/probe covers state All intracavity transducers should be covered with a single-use high-quality transducer cover. This may include some , but not all brands of condom, specific transducer covers or surgical drapes.
    Can you tell me how I determine which brands of condoms are suitable? Is there a criteria to assist?
    thanks

    Cate Coffey | Clinical Nurse Consultant
    Infection Prevention and Control Unit | Central Australia Health Service
    Northern Territory Government
    Alice Springs Hopsital, Gap Rd, Alice Springs
    GPO Box 2234, Suburb, NT Postcode
    p … 08 89517737
    e … cate.coffey@nt.gov.au http://www.nt.gov.au/health

    Our Vision: Better health outcomes for all Central Australians
    Our Values: Community at the Centre | Equity and Integrity | We are Accountable | We are Relevant Today and Ready for Tomorrow | We are Committed to High Quality Care | We Value our Partnerships

    Central Australia Health Service is a Smoke Free Workplace

    The information in this e-mail is intended solely for the addressee named. It may contain legally privileged or confidential information that is subject to copyright. If you are not the intended recipient you must not use, disclose copy or distribute this communication. If you have received this message in error, please delete the e-mail and notify the sender. No representation is made that this e-mail is free of viruses. Virus scanning is recommended and is the responsibility of the recipient.

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au
    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au
    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

Viewing 2 posts - 1 through 2 (of 2 total)
  • The forum ‘Infexion Connexion’ is closed to new topics and replies.