Lab accreditation for water testing?

Home Forums Infexion Connexion Lab accreditation for water testing?

 | Click to Receive Email Notifications of Posts
Viewing 9 posts - 1 through 9 (of 9 total)
  • Author
    Posts
  • #76114
    Michael Wishart
    Participant

    Author:
    Michael Wishart

    Position:

    Organisation:

    State:
    NSW

    Hi all

    Have a question. Like many of us, we have been microbiologically testing rinse waters from endoscopes and automated endoscope reprocesors (AERs) as per the GENCA/GESA guidelines for many years, via our human pathology testing lab. The lab has always maintained that it does these tests for us as a ‘favour’ (although we do pay… 🙂 ), as they are not NATA accredited to perform environmental testing, and that is what these tests actually are.

    Now that the final advisory for AS/NZS4187:2014 has been released from the ASCQHC, and there is a definite requirement to perform water testing for these AERs by December 2021, we also have to consider endotoxin testing (which our human pathology lab does not do), and possibly some chemical purity tests (dependent upon AER manufacturer advice). Now we should probably consider whether all of the final rinse water testing should be done by a lab accredited by NATA for this purpose (eg an environmental testing lab), despite some of the components of testing are already being done by our human pathology lab.

    AS/NZS4187: 2014 amendment 2:2019 has the following tests required for final rinse water in AERs:

    Total viable count 10 cfu/100 mL
    Pseudomonas aeruginosa Not detected/100 mL
    (Atypical) Mycobacterium sp Not detected/ 100 mL
    Chemical purity (as per manufacturer)
    Endotoxin 30 EU/mL

    Of those, we already routinely perform the first three with our human pathology testing accredited lab. Not sure what methods they currently use (the standard is very specific), so would need to check that as well.

    So, the question becomes: should we send ALL of our final rinse water specimens to an environmental testing accredited lab? Does anyone know if the NSQHS Standards accreditors will care which lab does what test, or they will look at the specific of the accreditation of the lab doing the testing, to measure compliance to this requirement of AS/NZS4187:2014?

    This is not an issue in CSSD as our human pathology lab has always declined to process those water samples, so we already have an environmental lab processing those specimens.

    Any thoughts or discussion would be valued. Am I being too pedantic about this, and no-one will care provided we get water tested?

    Thanks
    Michael

    Michael Wishart | Infection Control Coordinator, CICP-E

    St Vincent’s Private Hospital Northside | 627 Rode Road CHERMSIDE QLD 4032
    T +61 7 3326 3068 | F +61 7 3607 2226
    E michael.wishart@svha.org.au |
    W https://www.svphn.org.au

    [cid:image001.jpg@01D46C86.4CDB6090]
    [2019 conference email signature]

    ______________________________________________________________________
    This email and any attachments to it (the “Email”) is confidential and is for the use only of the intended recipient, and may not be duplicated or used by any other party without the express consent of the sender. If you are not the intended recipient of the Email, please notify the sender immediately by return email, delete the Email, and do not copy, print, retransmit, store or act in reliance on the Email. St Vincent’s Health Australia (“SVHA”) does not guarantee that the Email is free from errors, viruses or interference. Emails to and from SVHA or its related entities may be scanned and filtered in locations outside Australia.

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au

    To send a message to the list administrator send an email to admin@acipc.org.au

    You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au

    #76118
    Thomson, Rachel EA (THS)
    Participant

    Author:
    Thomson, Rachel EA (THS)

    Position:

    Organisation:

    State:

    Hi Michael,

    In response to your questions;

    The question becomes: should we send ALL of our final rinse water specimens to an environmental testing accredited lab?

    * We have changed our processes and hare now sending all of our specimens to a NATA approved laboratory

    Does anyone know if the NSQHS Standards accreditors will care which lab does what test, or they will look at the specific of the accreditation of the lab doing the testing, to measure compliance to this requirement of AS/NZS4187:2014

    * Uncertain of this, but our laboratory has recommended that we undertake all testing through a NATA certified laboratory.

    Kind regards
    Rachel

    ……………………………………………………………………………..
    Rachel Thomson
    Nurse Unit Manager

    Infection Prevention & Control Unit
    Royal Hobart Hospital
    Tasmanian Health Organisation-South

    *: 03 6166 7882/ 6166 8658

    Level 4, H Block
    48 Liverpool Street
    Hobart, 7000

    Hi all

    Have a question. Like many of us, we have been microbiologically testing rinse waters from endoscopes and automated endoscope reprocesors (AERs) as per the GENCA/GESA guidelines for many years, via our human pathology testing lab. The lab has always maintained that it does these tests for us as a ‘favour’ (although we do pay… 🙂 ), as they are not NATA accredited to perform environmental testing, and that is what these tests actually are.

    Now that the final advisory for AS/NZS4187:2014 has been released from the ASCQHC, and there is a definite requirement to perform water testing for these AERs by December 2021, we also have to consider endotoxin testing (which our human pathology lab does not do), and possibly some chemical purity tests (dependent upon AER manufacturer advice). Now we should probably consider whether all of the final rinse water testing should be done by a lab accredited by NATA for this purpose (eg an environmental testing lab), despite some of the components of testing are already being done by our human pathology lab.

    AS/NZS4187: 2014 amendment 2:2019 has the following tests required for final rinse water in AERs:

    Total viable count 10 cfu/100 mL
    Pseudomonas aeruginosa Not detected/100 mL
    (Atypical) Mycobacterium sp Not detected/ 100 mL
    Chemical purity (as per manufacturer)
    Endotoxin 30 EU/mL

    Of those, we already routinely perform the first three with our human pathology testing accredited lab. Not sure what methods they currently use (the standard is very specific), so would need to check that as well.

    So, the question becomes: should we send ALL of our final rinse water specimens to an environmental testing accredited lab? Does anyone know if the NSQHS Standards accreditors will care which lab does what test, or they will look at the specific of the accreditation of the lab doing the testing, to measure compliance to this requirement of AS/NZS4187:2014?

    This is not an issue in CSSD as our human pathology lab has always declined to process those water samples, so we already have an environmental lab processing those specimens.

    Any thoughts or discussion would be valued. Am I being too pedantic about this, and no-one will care provided we get water tested?

    Thanks
    Michael

    Michael Wishart | Infection Control Coordinator, CICP-E

    St Vincent’s Private Hospital Northside | 627 Rode Road CHERMSIDE QLD 4032
    T +61 7 3326 3068 | F +61 7 3607 2226
    E michael.wishart@svha.org.au |
    W https://www.svphn.org.au

    [cid:image001.jpg@01D46C86.4CDB6090]
    [2019 conference email signature]

    ______________________________________________________________________
    This email and any attachments to it (the “Email”) is confidential and is for the use only of the intended recipient, and may not be duplicated or used by any other party without the express consent of the sender. If you are not the intended recipient of the Email, please notify the sender immediately by return email, delete the Email, and do not copy, print, retransmit, store or act in reliance on the Email. St Vincent’s Health Australia (“SVHA”) does not guarantee that the Email is free from errors, viruses or interference. Emails to and from SVHA or its related entities may be scanned and filtered in locations outside Australia.
    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au

    To send a message to the list administrator send an email to admin@acipc.org.au

    You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au

    ________________________________

    CONFIDENTIALITY NOTICE AND DISCLAIMER
    The information in this transmission may be confidential and/or protected by legal professional privilege, and is intended only for the person or persons to whom it is addressed. If you are not such a person, you are warned that any disclosure, copying or dissemination of the information is unauthorised. If you have received the transmission in error, please immediately contact this office by telephone, fax or email, to inform us of the error and to enable arrangements to be made for the destruction of the transmission, or its return at our cost. No liability is accepted for any unauthorised use of the information contained in this transmission.

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au

    To send a message to the list administrator send an email to admin@acipc.org.au

    You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au

    #76132
    Anonymous
    Inactive

    Author:
    Anonymous

    Position:

    Organisation:

    State:

    Michael,

    I agree, I don’t think you are being pedantic at all. Normally everything is fine until there is a problem, i.e., you get an unexpected result. Then starts the usual cascade of events where you try to figure out why you have the result that you do. Is it a collection error, a testing error or and is it real?

    If you are not dealing with an accredited lab, this is going to make it challenging to progress. Another issue that I have recently come across, is that it is not only an accredited lab that matters, but also that the test performed is NATA accredited as well AND to the standard specified.

    My recent experience is with the requirement for Atypical mycobacteria testing or NTM. There are not many labs that actually perform this test as per the recommended methodology set out in ISO 15883-4 (see Annex E for detail). The water samples were sent to an accredited Environmental Laboratory, however as I was reviewing the results, I happened to read the fine print of the report and discovered that this was not a NATA accredited test. Then when we got a positive result, it became “challenging” to know what action to take, as there is always pressure to continue to use the equipment.

    As a side point, because the decision usually on which lab to use is often a financial one, I recently received some great advice. If you are going to use a lab that is not accredited, then ensure that it is documented in your organisations risk register. I would also include actions to take when you do have a positive result.

    Have a great day and good-luck.

    Mandy Davidson
    RN; GCert Inf Pre & Cont; MPHTM; Cert III Sterilisation; Cert IV TAE; Immunisation cred; CICP-A
    Clinical Nurse Consultant – 4187 Implementation project
    Infection Prevention & Control

    [cid:image001.png@01D3A192.E1513890]
    T
    07 4433 1873 | 0402 987 432
    E
    Mandy.Davidson@health.qld.gov.au
    W
    http://www.health.qld.gov.au/townsville

    Townsville Hospital and Health Service
    100 Angus Smith Drive, Douglas, QLD 4814
    [Facebook-Icon] [Twitter-Icon] [Linkedin-Icon]
    Townsville Hospital and Health Service acknowledges the Traditional Owners of the land, and pays respect to Elders past, present and future.

    Hi all

    Have a question. Like many of us, we have been microbiologically testing rinse waters from endoscopes and automated endoscope reprocesors (AERs) as per the GENCA/GESA guidelines for many years, via our human pathology testing lab. The lab has always maintained that it does these tests for us as a ‘favour’ (although we do pay… 🙂 ), as they are not NATA accredited to perform environmental testing, and that is what these tests actually are.

    Now that the final advisory for AS/NZS4187:2014 has been released from the ASCQHC, and there is a definite requirement to perform water testing for these AERs by December 2021, we also have to consider endotoxin testing (which our human pathology lab does not do), and possibly some chemical purity tests (dependent upon AER manufacturer advice). Now we should probably consider whether all of the final rinse water testing should be done by a lab accredited by NATA for this purpose (eg an environmental testing lab), despite some of the components of testing are already being done by our human pathology lab.

    AS/NZS4187: 2014 amendment 2:2019 has the following tests required for final rinse water in AERs:

    Total viable count 10 cfu/100 mL
    Pseudomonas aeruginosa Not detected/100 mL
    (Atypical) Mycobacterium sp Not detected/ 100 mL
    Chemical purity (as per manufacturer)
    Endotoxin 30 EU/mL

    Of those, we already routinely perform the first three with our human pathology testing accredited lab. Not sure what methods they currently use (the standard is very specific), so would need to check that as well.

    So, the question becomes: should we send ALL of our final rinse water specimens to an environmental testing accredited lab? Does anyone know if the NSQHS Standards accreditors will care which lab does what test, or they will look at the specific of the accreditation of the lab doing the testing, to measure compliance to this requirement of AS/NZS4187:2014?

    This is not an issue in CSSD as our human pathology lab has always declined to process those water samples, so we already have an environmental lab processing those specimens.

    Any thoughts or discussion would be valued. Am I being too pedantic about this, and no-one will care provided we get water tested?

    Thanks
    Michael

    Michael Wishart | Infection Control Coordinator, CICP-E

    St Vincent’s Private Hospital Northside | 627 Rode Road CHERMSIDE QLD 4032
    T +61 7 3326 3068 | F +61 7 3607 2226
    E michael.wishart@svha.org.au |
    W https://www.svphn.org.au

    [cid:image001.jpg@01D46C86.4CDB6090]
    [2019 conference email signature]

    ______________________________________________________________________
    This email and any attachments to it (the “Email”) is confidential and is for the use only of the intended recipient, and may not be duplicated or used by any other party without the express consent of the sender. If you are not the intended recipient of the Email, please notify the sender immediately by return email, delete the Email, and do not copy, print, retransmit, store or act in reliance on the Email. St Vincent’s Health Australia (“SVHA”) does not guarantee that the Email is free from errors, viruses or interference. Emails to and from SVHA or its related entities may be scanned and filtered in locations outside Australia.
    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au

    To send a message to the list administrator send an email to admin@acipc.org.au

    You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au

    ********************************************************************************

    This email, including any attachments sent with it, is confidential and for the sole use of the intended recipient(s). This confidentiality is not waived or lost, if you receive it and you are not the intended recipient(s), or if it is transmitted/received in error.

    Any unauthorised use, alteration, disclosure, distribution or review of this email is strictly prohibited. The information contained in this email, including any attachment sent with it, may be subject to a statutory duty of confidentiality if it relates to health service matters.

    If you are not the intended recipient(s), or if you have received this email in error, you are asked to immediately notify the sender by telephone collect on Australia +61 1800 198 175 or by return email. You should also delete this email, and any copies, from your computer system network and destroy any hard copies produced.

    If not an intended recipient of this email, you must not copy, distribute or take any action(s) that relies on it; any form of disclosure, modification, distribution and/or publication of this email is also prohibited.

    Although Queensland Health takes all reasonable steps to ensure this email does not contain malicious software, Queensland Health does not accept responsibility for the consequences if any person’s computer inadvertently suffers any disruption to services, loss of information, harm or is infected with a virus, other malicious computer programme or code that may occur as a consequence of receiving this email.

    Unless stated otherwise, this email represents only the views of the sender and not the views of the Queensland Government.

    **********************************************************************************

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au

    To send a message to the list administrator send an email to admin@acipc.org.au

    You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au

    #76135
    Donna Cameron
    Participant

    Author:
    Donna Cameron

    Position:
    Infection Control Consultant

    Organisation:
    University of Melbourne

    State:
    VIC

    #76136
    Michael Wishart
    Participant

    Author:
    Michael Wishart

    Position:

    Organisation:

    State:
    NSW

    [Posted on behalf of Dr Robson, with thanks – Moderator]

    Hi Michael et al
    Thought I would pass on that there is limited NATA accreditation available for Human Pathology labs in the environmental testing arena (see page 10 of 15 in the SoA).
    For example our lab has NATA accreditation for environmental testing as below. It does not however fulfil all the requirements of AS/NZS4187:2014 (endotoxin testing).

    [cid:image001.png@01D5C475.8F43A350]

    Additionally the ACQSHC and MTAC recommendations for Heater Cooler Unit mycobacterial testing (but not hetertrophic bacterial counts) can only be performed in Human Pathology Clinical Lab. The culture requirements for non-rapidly growing mycobacteria (e.g chimera) cannot usually be performed in environmental testing laboratories.

    Currently there is an RCPAQAP module available for Human pathology laboratory laboratories to perform endoscope testing and moves afoot to also possibly introduce a QAP for Mycobacterial testing to satisfy heater cooler unit requirements. To have NATA accreditation for a particular test the laboratory is required to be enrolled in a formal QAP programme or if not available a laboratory exchange programme.

    Kind regards
    Dr Jenny Robson
    Sullivan Nicolaides Pathology

    Michael,

    I agree, I don’t think you are being pedantic at all. Normally everything is fine until there is a problem, i.e., you get an unexpected result. Then starts the usual cascade of events where you try to figure out why you have the result that you do. Is it a collection error, a testing error or and is it real?

    If you are not dealing with an accredited lab, this is going to make it challenging to progress. Another issue that I have recently come across, is that it is not only an accredited lab that matters, but also that the test performed is NATA accredited as well AND to the standard specified.

    My recent experience is with the requirement for Atypical mycobacteria testing or NTM. There are not many labs that actually perform this test as per the recommended methodology set out in ISO 15883-4 (see Annex E for detail). The water samples were sent to an accredited Environmental Laboratory, however as I was reviewing the results, I happened to read the fine print of the report and discovered that this was not a NATA accredited test. Then when we got a positive result, it became “challenging” to know what action to take, as there is always pressure to continue to use the equipment.

    As a side point, because the decision usually on which lab to use is often a financial one, I recently received some great advice. If you are going to use a lab that is not accredited, then ensure that it is documented in your organisations risk register. I would also include actions to take when you do have a positive result.

    Have a great day and good-luck.

    Mandy Davidson
    RN; GCert Inf Pre & Cont; MPHTM; Cert III Sterilisation; Cert IV TAE; Immunisation cred; CICP-A
    Clinical Nurse Consultant – 4187 Implementation project
    Infection Prevention & Control

    [cid:image001.png@01D3A192.E1513890]
    T
    07 4433 1873 | 0402 987 432
    E
    Mandy.Davidson@health.qld.gov.au
    W
    http://www.health.qld.gov.au/townsville

    Townsville Hospital and Health Service
    100 Angus Smith Drive, Douglas, QLD 4814
    [Facebook-Icon] [Twitter-Icon] [Linkedin-Icon]
    Townsville Hospital and Health Service acknowledges the Traditional Owners of the land, and pays respect to Elders past, present and future.

    Hi all

    Have a question. Like many of us, we have been microbiologically testing rinse waters from endoscopes and automated endoscope reprocesors (AERs) as per the GENCA/GESA guidelines for many years, via our human pathology testing lab. The lab has always maintained that it does these tests for us as a ‘favour’ (although we do pay… 🙂 ), as they are not NATA accredited to perform environmental testing, and that is what these tests actually are.

    Now that the final advisory for AS/NZS4187:2014 has been released from the ASCQHC, and there is a definite requirement to perform water testing for these AERs by December 2021, we also have to consider endotoxin testing (which our human pathology lab does not do), and possibly some chemical purity tests (dependent upon AER manufacturer advice). Now we should probably consider whether all of the final rinse water testing should be done by a lab accredited by NATA for this purpose (eg an environmental testing lab), despite some of the components of testing are already being done by our human pathology lab.

    AS/NZS4187: 2014 amendment 2:2019 has the following tests required for final rinse water in AERs:

    Total viable count 10 cfu/100 mL
    Pseudomonas aeruginosa Not detected/100 mL
    (Atypical) Mycobacterium sp Not detected/ 100 mL
    Chemical purity (as per manufacturer)
    Endotoxin 30 EU/mL

    Of those, we already routinely perform the first three with our human pathology testing accredited lab. Not sure what methods they currently use (the standard is very specific), so would need to check that as well.

    So, the question becomes: should we send ALL of our final rinse water specimens to an environmental testing accredited lab? Does anyone know if the NSQHS Standards accreditors will care which lab does what test, or they will look at the specific of the accreditation of the lab doing the testing, to measure compliance to this requirement of AS/NZS4187:2014?

    This is not an issue in CSSD as our human pathology lab has always declined to process those water samples, so we already have an environmental lab processing those specimens.

    Any thoughts or discussion would be valued. Am I being too pedantic about this, and no-one will care provided we get water tested?

    Thanks
    Michael

    Michael Wishart | Infection Control Coordinator, CICP-E

    St Vincent’s Private Hospital Northside | 627 Rode Road CHERMSIDE QLD 4032
    T +61 7 3326 3068 | F +61 7 3607 2226
    E michael.wishart@svha.org.au |
    W https://www.svphn.org.au

    [cid:image001.jpg@01D46C86.4CDB6090]
    [2019 conference email signature]

    ______________________________________________________________________

    ______________________________________________________________________
    This email and any attachments to it (the “Email”) is confidential and is for the use only of the intended recipient, and may not be duplicated or used by any other party without the express consent of the sender. If you are not the intended recipient of the Email, please notify the sender immediately by return email, delete the Email, and do not copy, print, retransmit, store or act in reliance on the Email. St Vincent’s Health Australia (“SVHA”) does not guarantee that the Email is free from errors, viruses or interference. Emails to and from SVHA or its related entities may be scanned and filtered in locations outside Australia.

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au

    To send a message to the list administrator send an email to admin@acipc.org.au

    You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au

    #76139
    Mona Schousboe
    Participant

    Author:
    Mona Schousboe

    Position:

    Organisation:

    State:

    I agree with Jenny Robson that the issue of environmental testing is not
    black or white. For testing of endoscopes, there are good and regularly
    updated guidelines. What was the history behind the requirements for
    endotoxin testing, and what was the research into the general availability
    of the test? While mycobacterial testing already is concentrated in
    specialist TB laboratories, atypical mycobacteria like *M marinum *is
    isolated on the general microbiology benches. Research into new pathogens
    like was required for Heater-Cooler Units water analyses was performed in
    TB laboratories, not in environmental laboratories. We have not even
    touched on the problems of organisms resistant to metals such as copper. We
    found the M chimaera was isolated from a Heater-Cooler Unit sample with a
    very high concentration of copper. We started environment testing for
    legionella pneumophila in hospital hot water systems in the 1980es- and
    published results many times. The requirements for environmental testing
    accreditation, with the associated high cost, was not because we could not
    culture for environmental * Legionella, *but because of IANZ (the NZ
    equivalent of NATA) require different programmes for water testing than the
    direct human specimen. They have no plan for swabs from showerheads. I
    wonder if the accreditation authorities should be more conscious of the
    cost of accreditation and become more versatile in the programmes they
    offer to Microbiology Laboratories.
    Kind regards
    Mona Schousboe
    FRCPA, MPH
    Christchurch

    On Mon, Jan 6, 2020 at 12:55 PM Michael Wishart
    wrote:

    > [Posted on behalf of Dr Robson, with thanks Moderator]
    >
    >
    >
    > Hi Michael et al
    >
    > Thought I would pass on that there is limited NATA accreditation available
    > for Human Pathology labs in the environmental testing arena (see page 10 of
    > 15 in the SoA).
    >
    > For example our lab has NATA accreditation for environmental testing as
    > below. It does not however fulfil all the requirements of AS/NZS4187:2014
    > (endotoxin testing).
    >
    >
    >
    >
    >
    > Additionally the ACQSHC and MTAC recommendations for Heater Cooler Unit
    > mycobacterial testing (but not hetertrophic bacterial counts) can only be
    > performed in Human Pathology Clinical Lab. The culture requirements for
    > non-rapidly growing mycobacteria (e.g chimera) cannot usually be performed
    > in environmental testing laboratories.
    >
    >
    >
    > Currently there is an RCPAQAP module available for Human pathology
    > laboratory laboratories to perform endoscope testing and moves afoot to
    > also possibly introduce a QAP for Mycobacterial testing to satisfy heater
    > cooler unit requirements. To have NATA accreditation for a particular test
    > the laboratory is required to be enrolled in a formal QAP programme or if
    > not available a laboratory exchange programme.
    >
    >
    >
    > Kind regards
    >
    > Dr Jenny Robson
    >
    > Sullivan Nicolaides Pathology
    >
    > *From:* ACIPC Infexion Connexion [mailto:ACIPCLIST@ACIPC.ORG.AU
    > ] *On Behalf Of *Mandy Davidson
    > *Sent:* Monday, 6 January 2020 8:14 AM
    > *To:* ACIPCLIST@ACIPC.ORG.AU
    > *Subject:* [External] Re: [ACIPC_Infexion_Connexion] Lab accreditation
    > for water testing?
    >
    >
    >
    > Michael,
    >
    >
    >
    > I agree, I dont think you are being pedantic at all. Normally everything
    > is fine until there is a problem, i.e., you get an unexpected result. Then
    > starts the usual cascade of events where you try to figure out why you have
    > the result that you do. Is it a collection error, a testing error or and
    > is it real?
    >
    >
    >
    > If you are not dealing with an accredited lab, this is going to make it
    > challenging to progress. Another issue that I have recently come across, is
    > that it is not only an accredited lab that matters, but also that the test
    > performed is NATA accredited as well AND to the standard specified.
    >
    >
    >
    > My recent experience is with the requirement for Atypical mycobacteria
    > testing or NTM. There are not many labs that actually perform this test as
    > per the recommended methodology set out in ISO 15883-4 (see Annex E for
    > detail). The water samples were sent to an accredited Environmental
    > Laboratory, however as I was reviewing the results, I happened to read the
    > fine print of the report and discovered that this was not a NATA accredited
    > test. Then when we got a positive result, it became challenging to know
    > what action to take, as there is always pressure to continue to use the
    > equipment.
    >
    >
    >
    > As a side point, because the decision usually on which lab to use is often
    > a financial one, I recently received some great advice. If you are going to
    > use a lab that is not accredited, then ensure that it is documented in your
    > organisations risk register. I would also include actions to take when you
    > do have a positive result.
    >
    >
    >
    > Have a great day and good-luck.
    >
    >
    >
    > *Mandy Davidson*
    >
    > *RN; GCert Inf Pre & Cont; MPHTM; Cert III Sterilisation; Cert IV TAE;
    > Immunisation cred; CICP-A*
    >
    > Clinical Nurse Consultant 4187 Implementation project
    >
    > Infection Prevention & Control
    >
    >
    >
    > [image: cid:image001.png@01D3A192.E1513890]
    >
    > *T*
    >
    > *07 4433 1873* | *0402 987 432*
    >
    > *E*
    >
    > Mandy.Davidson@health.qld.gov.au
    >
    > *W*
    >
    > http://www.health.qld.gov.au/townsville
    >
    >
    >
    >
    >
    > *Townsville Hospital and Health Service*
    >
    > 100 Angus Smith Drive, Douglas, QLD 4814
    >
    > *[image: Facebook-Icon]*
    >
    > *[image: Twitter-Icon]*
    >
    > *[image: Linkedin-Icon]*
    >
    >
    > *Townsville Hospital and Health Service acknowledges the Traditional
    > Owners of the land, and pays respect to Elders past, present and future.*
    >
    >
    >
    >
    >
    > *From:* ACIPC Infexion Connexion *On Behalf Of *Michael
    > Wishart
    > *Sent:* Thursday, 19 December 2019 11:51 AM
    > *To:* ACIPCLIST@ACIPC.ORG.AU
    > *Subject:* [ACIPC_Infexion_Connexion] Lab accreditation for water testing?
    >
    >
    >
    > Hi all
    >
    >
    >
    > Have a question. Like many of us, we have been microbiologically testing
    > rinse waters from endoscopes and automated endoscope reprocesors (AERs) as
    > per the GENCA/GESA guidelines for many years, via our human pathology
    > testing lab. The lab has always maintained that it does these tests for us
    > as a favour (although we do pay J ), as they are not NATA accredited
    > to perform environmental testing, and that is what these tests actually are.
    >
    >
    >
    > Now that the final advisory for AS/NZS4187:2014 has been released from the
    > ASCQHC, and there is a definite requirement to perform water testing for
    > these AERs by December 2021, we also have to consider endotoxin testing
    > (which our human pathology lab does not do), and possibly some chemical
    > purity tests (dependent upon AER manufacturer advice). Now we should
    > probably consider whether all of the final rinse water testing should be
    > done by a lab accredited by NATA for this purpose (eg an environmental
    > testing lab), despite some of the components of testing are already being
    > done by our human pathology lab.
    >
    >
    >
    > AS/NZS4187: 2014 amendment 2:2019 has the following tests required for
    > final rinse water in AERs:
    >
    >
    >
    > Total viable count 10 cfu/100 mL
    >
    > Pseudomonas aeruginosa Not detected/100 mL
    >
    > (Atypical) Mycobacterium sp Not detected/ 100 mL
    >
    > Chemical purity (as per manufacturer)
    >
    > Endotoxin 30 EU/mL
    >
    >
    >
    > Of those, we already routinely perform the first three with our human
    > pathology testing accredited lab. Not sure what methods they currently use
    > (the standard is very specific), so would need to check that as well.
    >
    >
    >
    > So, the question becomes: should we send ALL of our final rinse water
    > specimens to an environmental testing accredited lab? Does anyone know if
    > the NSQHS Standards accreditors will care which lab does what test, or they
    > will look at the specific of the accreditation of the lab doing the
    > testing, to measure compliance to this requirement of AS/NZS4187:2014?
    >
    >
    >
    > This is not an issue in CSSD as our human pathology lab has always
    > declined to process those water samples, so we already have an
    > environmental lab processing those specimens.
    >
    >
    >
    > Any thoughts or discussion would be valued. Am I being too pedantic about
    > this, and no-one will care provided we get water tested?
    >
    >
    >
    > Thanks
    >
    > Michael
    >
    >
    >
    >
    >
    > *Michael Wishart *| Infection Control Coordinator, CICP-E
    >
    >
    > St Vincents Private Hospital Northside | 627 Rode Road CHERMSIDE QLD
    > 4032
    >
    > *T *+61 7 3326 3068 |* F* +61 7 3607 2226
    >
    > *E* michael.wishart@svha.org.au |
    >
    > *W *https://www.svphn.org.au
    >
    >
    >
    >
    > [image: cid:image001.jpg@01D46C86.4CDB6090]
    >
    > [image: 2019 conference email signature]
    >
    >
    >
    >
    > ______________________________________________________________________
    >
    >
    > ______________________________________________________________________
    > This email and any attachments to it (the “Email”) is confidential and is
    > for the use only of the intended recipient, and may not be duplicated or
    > used by any other party without the express consent of the sender. If you
    > are not the intended recipient of the Email, please notify the sender
    > immediately by return email, delete the Email, and do not copy, print,
    > retransmit, store or act in reliance on the Email. St Vincent’s Health
    > Australia (“SVHA”) does not guarantee that the Email is free from errors,
    > viruses or interference. Emails to and from SVHA or its related entities
    > may be scanned and filtered in locations outside Australia.
    > MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO
    > NOT REPRESENT THE OPINION OF ACIPC.
    >
    > The use of trade/product/commercial brand names through the list is
    > discouraged by ACIPC. If you wish to discuss specific reference to products
    > or services by brand or commercial names, please do this outside the list.
    >
    > Archive of all messages are available at http://aicalist.org.au/archives
    > – registration and login required.
    >
    > Replies to this message will be directed back to the list. To create a new
    > message send an email to acipclist@acipc.org.au
    >
    > To send a message to the list administrator send an email to
    > admin@acipc.org.au
    >
    > You can unsubscribe manually from this list by sending ‘signoff acipclist’
    > (without the quotes) to listserv@aicalist.org.au
    >

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au

    To send a message to the list administrator send an email to admin@acipc.org.au

    You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au

    #76140
    Michael Wishart
    Participant

    Author:
    Michael Wishart

    Position:

    Organisation:

    State:
    NSW

    Thanks to Jenny Robson and Mona Schousboe for the lab perspective in this. And thanks to all others who have agreed this could be an issue we need to be aware of.

    We have two separate types of microbiological testing here, required by two different agencies.

    1. Endoscope testing (as a QA activity to ensure cleaning is appropriate) is required by GESA/GENCA Infection Control in Endoscopy 2010 (havent yet seen the draft revision, has anybody?), and includes testing for CRE post use on known CRE patients (as per 2017 consensus guidelines). Endoscope microbiological testing is not mentioned in AS/NZS 4187 per se (only as general part of a review of storage of RMDs I would suggest).

    2. Testing of the rinse water for AERs is recommended by both, but GESA/GENCA only refer to microbiological monitoring of rinse water to ensure sterility, whereas AS/NZS 4187 looks at rinse water with regard to its suitability to contact RMDs, not just microbial content (hence the extra testing required), similar to that required in CSSD. I am sure Standards Australia have some good references for the need for endotoxin testing in these environments.

    To me, it looks like, as long as your path lab follows appropriate processes for testing (as per the required standards) and is accredited for the type of testing required, it should be fine. It would be prudent to check what tests they are using, and ensure these are as the relevant standard or guideline requires.

    Both Jenny and Mona mentioned testing for M. chimaera in cardiac heater-coolers, and this is another example of the difficulties in getting testing done appropriately, as I believe most environmental labs have difficulty isolating M. chimaera specifically.

    It would be nice to have some consistency of guidelines for testing in endoscopy, so hopefully the revision of the GESA/GENCA Infection Control in Endoscopy document will help with this, not create even more confusion!

    Cheers
    Michael

    Michael Wishart | Infection Control Coordinator, CICP-E

    St Vincents Private Hospital Northside | 627 Rode Road CHERMSIDE QLD 4032
    T +61 7 3326 3068 | F +61 7 3607 2226
    E michael.wishart@svha.org.au |
    W https://www.svphn.org.au

    [cid:image001.jpg@01D46C86.4CDB6090]

    From: ACIPC Infexion Connexion On Behalf Of Mona Schousboe
    Sent: Monday, 6 January 2020 11:52 AM
    To: ACIPCLIST@ACIPC.ORG.AU
    Subject: Re: [ACIPC_Infexion_Connexion] Lab accreditation for water testing?

    I agree with Jenny Robson that the issue of environmental testing is not black or white. For testing of endoscopes, there are good and regularly updated guidelines. What was the history behind the requirements for endotoxin testing, and what was the research into the general availability of the test? While mycobacterial testing already is concentrated in specialist TB laboratories, atypical mycobacteria like M marinum is isolated on the general microbiology benches. Research into new pathogens like was required for Heater-Cooler Units water analyses was performed in TB laboratories, not in environmental laboratories. We have not even touched on the problems of organisms resistant to metals such as copper. We found the M chimaera was isolated from a Heater-Cooler Unit sample with a very high concentration of copper. We started environment testing for legionella pneumophila in hospital hot water systems in the 1980es- and published results many times. The requirements for environmental testing accreditation, with the associated high cost, was not because we could not culture for environmental Legionella, but because of IANZ (the NZ equivalent of NATA) require different programmes for water testing than the direct human specimen. They have no plan for swabs from showerheads. I wonder if the accreditation authorities should be more conscious of the cost of accreditation and become more versatile in the programmes they offer to Microbiology Laboratories.
    Kind regards
    Mona Schousboe
    FRCPA, MPH
    Christchurch

    On Mon, Jan 6, 2020 at 12:55 PM Michael Wishart <Michael.Wishart@svha.org.au> wrote:
    [Posted on behalf of Dr Robson, with thanks Moderator]

    Hi Michael et al
    Thought I would pass on that there is limited NATA accreditation available for Human Pathology labs in the environmental testing arena (see page 10 of 15 in the SoA).
    For example our lab has NATA accreditation for environmental testing as below. It does not however fulfil all the requirements of AS/NZS4187:2014 (endotoxin testing).

    Additionally the ACQSHC and MTAC recommendations for Heater Cooler Unit mycobacterial testing (but not hetertrophic bacterial counts) can only be performed in Human Pathology Clinical Lab. The culture requirements for non-rapidly growing mycobacteria (e.g chimera) cannot usually be performed in environmental testing laboratories.

    Currently there is an RCPAQAP module available for Human pathology laboratory laboratories to perform endoscope testing and moves afoot to also possibly introduce a QAP for Mycobacterial testing to satisfy heater cooler unit requirements. To have NATA accreditation for a particular test the laboratory is required to be enrolled in a formal QAP programme or if not available a laboratory exchange programme.

    Kind regards
    Dr Jenny Robson
    Sullivan Nicolaides Pathology
    From: ACIPC Infexion Connexion [mailto:ACIPCLIST@ACIPC.ORG.AU] On Behalf Of Mandy Davidson
    Sent: Monday, 6 January 2020 8:14 AM
    To: ACIPCLIST@ACIPC.ORG.AU
    Subject: [External] Re: [ACIPC_Infexion_Connexion] Lab accreditation for water testing?

    Michael,

    I agree, I dont think you are being pedantic at all. Normally everything is fine until there is a problem, i.e., you get an unexpected result. Then starts the usual cascade of events where you try to figure out why you have the result that you do. Is it a collection error, a testing error or and is it real?

    If you are not dealing with an accredited lab, this is going to make it challenging to progress. Another issue that I have recently come across, is that it is not only an accredited lab that matters, but also that the test performed is NATA accredited as well AND to the standard specified.

    My recent experience is with the requirement for Atypical mycobacteria testing or NTM. There are not many labs that actually perform this test as per the recommended methodology set out in ISO 15883-4 (see Annex E for detail). The water samples were sent to an accredited Environmental Laboratory, however as I was reviewing the results, I happened to read the fine print of the report and discovered that this was not a NATA accredited test. Then when we got a positive result, it became challenging to know what action to take, as there is always pressure to continue to use the equipment.

    As a side point, because the decision usually on which lab to use is often a financial one, I recently received some great advice. If you are going to use a lab that is not accredited, then ensure that it is documented in your organisations risk register. I would also include actions to take when you do have a positive result.

    Have a great day and good-luck.

    Mandy Davidson
    RN; GCert Inf Pre & Cont; MPHTM; Cert III Sterilisation; Cert IV TAE; Immunisation cred; CICP-A
    Clinical Nurse Consultant 4187 Implementation project
    Infection Prevention & Control

    T
    07 4433 1873 | 0402 987 432
    E
    Mandy.Davidson@health.qld.gov.au
    W
    http://www.health.qld.gov.au/townsville

    Townsville Hospital and Health Service
    100 Angus Smith Drive, Douglas, QLD 4814

    Townsville Hospital and Health Service acknowledges the Traditional Owners of the land, and pays respect to Elders past, present and future.

    From: ACIPC Infexion Connexion On Behalf Of Michael Wishart
    Sent: Thursday, 19 December 2019 11:51 AM
    To: ACIPCLIST@ACIPC.ORG.AU
    Subject: [ACIPC_Infexion_Connexion] Lab accreditation for water testing?

    Hi all

    Have a question. Like many of us, we have been microbiologically testing rinse waters from endoscopes and automated endoscope reprocesors (AERs) as per the GENCA/GESA guidelines for many years, via our human pathology testing lab. The lab has always maintained that it does these tests for us as a favour (although we do pay ), as they are not NATA accredited to perform environmental testing, and that is what these tests actually are.

    Now that the final advisory for AS/NZS4187:2014 has been released from the ASCQHC, and there is a definite requirement to perform water testing for these AERs by December 2021, we also have to consider endotoxin testing (which our human pathology lab does not do), and possibly some chemical purity tests (dependent upon AER manufacturer advice). Now we should probably consider whether all of the final rinse water testing should be done by a lab accredited by NATA for this purpose (eg an environmental testing lab), despite some of the components of testing are already being done by our human pathology lab.

    AS/NZS4187: 2014 amendment 2:2019 has the following tests required for final rinse water in AERs:

    Total viable count 10 cfu/100 mL
    Pseudomonas aeruginosa Not detected/100 mL
    (Atypical) Mycobacterium sp Not detected/ 100 mL
    Chemical purity (as per manufacturer)
    Endotoxin 30 EU/mL

    Of those, we already routinely perform the first three with our human pathology testing accredited lab. Not sure what methods they currently use (the standard is very specific), so would need to check that as well.

    So, the question becomes: should we send ALL of our final rinse water specimens to an environmental testing accredited lab? Does anyone know if the NSQHS Standards accreditors will care which lab does what test, or they will look at the specific of the accreditation of the lab doing the testing, to measure compliance to this requirement of AS/NZS4187:2014?

    This is not an issue in CSSD as our human pathology lab has always declined to process those water samples, so we already have an environmental lab processing those specimens.

    Any thoughts or discussion would be valued. Am I being too pedantic about this, and no-one will care provided we get water tested?

    Thanks
    Michael

    Michael Wishart | Infection Control Coordinator, CICP-E

    St Vincents Private Hospital Northside | 627 Rode Road CHERMSIDE QLD 4032
    T +61 7 3326 3068 | F +61 7 3607 2226
    E michael.wishart@svha.org.au |
    W https://www.svphn.org.au

    ______________________________________________________________________

    ______________________________________________________________________
    This email and any attachments to it (the “Email”) is confidential and is for the use only of the intended recipient, and may not be duplicated or used by any other party without the express consent of the sender. If you are not the intended recipient of the Email, please notify the sender immediately by return email, delete the Email, and do not copy, print, retransmit, store or act in reliance on the Email. St Vincent’s Health Australia (“SVHA”) does not guarantee that the Email is free from errors, viruses or interference. Emails to and from SVHA or its related entities may be scanned and filtered in locations outside Australia.
    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au

    To send a message to the list administrator send an email to admin@acipc.org.au

    You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au
    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au

    To send a message to the list administrator send an email to admin@acipc.org.au

    You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au
    ______________________________________________________________________
    This email and any attachments to it (the “Email”) is confidential and is for the use only of the intended recipient, and may not be duplicated or used by any other party without the express consent of the sender. If you are not the intended recipient of the Email, please notify the sender immediately by return email, delete the Email, and do not copy, print, retransmit, store or act in reliance on the Email. St Vincent’s Health Australia (“SVHA”) does not guarantee that the Email is free from errors, viruses or interference. Emails to and from SVHA or its related entities may be scanned and filtered in locations outside Australia.

    ______________________________________________________________________
    This email and any attachments to it (the “Email”) is confidential and is for the use only of the intended recipient, and may not be duplicated or used by any other party without the express consent of the sender. If you are not the intended recipient of the Email, please notify the sender immediately by return email, delete the Email, and do not copy, print, retransmit, store or act in reliance on the Email. St Vincent’s Health Australia (“SVHA”) does not guarantee that the Email is free from errors, viruses or interference. Emails to and from SVHA or its related entities may be scanned and filtered in locations outside Australia.

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au

    To send a message to the list administrator send an email to admin@acipc.org.au

    You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quo

    #76141
    Morris, Sylvia (Health)
    Participant

    Author:
    Morris, Sylvia (Health)

    Position:

    Organisation:

    State:

    #76154
    Karen Vickery
    Participant

    Author:
    Karen Vickery

    Position:

    Organisation:

    State:

    HI Michael,
    Re repeated sterilisation of orthopaedic screws. Could you please forward on that soil accumulation and biofilm formation is also a risk. Please see the sem in attached paper. Figure 4

    Regards
    Karen Vickery BVSc, MVSc, PhD
    Macquarie University
    Karen.vickery@mq.edu.au

    Hi all

    Have a question. Like many of us, we have been microbiologically testing rinse waters from endoscopes and automated endoscope reprocesors (AERs) as per the GENCA/GESA guidelines for many years, via our human pathology testing lab. The lab has always maintained that it does these tests for us as a ‘favour’ (although we do pay… 🙂 ), as they are not NATA accredited to perform environmental testing, and that is what these tests actually are.

    Now that the final advisory for AS/NZS4187:2014 has been released from the ASCQHC, and there is a definite requirement to perform water testing for these AERs by December 2021, we also have to consider endotoxin testing (which our human pathology lab does not do), and possibly some chemical purity tests (dependent upon AER manufacturer advice). Now we should probably consider whether all of the final rinse water testing should be done by a lab accredited by NATA for this purpose (eg an environmental testing lab), despite some of the components of testing are already being done by our human pathology lab.

    AS/NZS4187: 2014 amendment 2:2019 has the following tests required for final rinse water in AERs:

    Total viable count 10 cfu/100 mL
    Pseudomonas aeruginosa Not detected/100 mL
    (Atypical) Mycobacterium sp Not detected/ 100 mL
    Chemical purity (as per manufacturer)
    Endotoxin 30 EU/mL

    Of those, we already routinely perform the first three with our human pathology testing accredited lab. Not sure what methods they currently use (the standard is very specific), so would need to check that as well.

    So, the question becomes: should we send ALL of our final rinse water specimens to an environmental testing accredited lab? Does anyone know if the NSQHS Standards accreditors will care which lab does what test, or they will look at the specific of the accreditation of the lab doing the testing, to measure compliance to this requirement of AS/NZS4187:2014?

    This is not an issue in CSSD as our human pathology lab has always declined to process those water samples, so we already have an environmental lab processing those specimens.

    Any thoughts or discussion would be valued. Am I being too pedantic about this, and no-one will care provided we get water tested?

    Thanks
    Michael

    Michael Wishart | Infection Control Coordinator, CICP-E

    St Vincent’s Private Hospital Northside | 627 Rode Road CHERMSIDE QLD 4032
    T +61 7 3326 3068 | F +61 7 3607 2226
    E michael.wishart@svha.org.au |
    W https://www.svphn.org.au

    [cid:image001.jpg@01D46C86.4CDB6090]
    [2019 conference email signature]

    ______________________________________________________________________
    This email and any attachments to it (the “Email”) is confidential and is for the use only of the intended recipient, and may not be duplicated or used by any other party without the express consent of the sender. If you are not the intended recipient of the Email, please notify the sender immediately by return email, delete the Email, and do not copy, print, retransmit, store or act in reliance on the Email. St Vincent’s Health Australia (“SVHA”) does not guarantee that the Email is free from errors, viruses or interference. Emails to and from SVHA or its related entities may be scanned and filtered in locations outside Australia.
    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au

    To send a message to the list administrator send an email to admin@acipc.org.au

    You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au

    To send a message to the list administrator send an email to admin@acipc.org.au

    You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au

Viewing 9 posts - 1 through 9 (of 9 total)
  • The forum ‘Infexion Connexion’ is closed to new topics and replies.