    Dear all,

    Find below the FDA notification of the New Reprocessing Instructions for the
    Olympus TJF-Q180V duodenoscopes.

    Attached is the Olympus letter to customers which details the new
    reprocessing instructions including;

    . supplementary flushing instructions

    . additional recess flushing and forceps elevator raising /lowering
    steps during pre-cleaning

    . manual cleaning

    . manual disinfection and endoscope rinsing and alcohol flushing.

    Olympus advise that “these new reprocessing procedures should be implemented
    as soon as possible” and that “the new cleaning procedure requires the use
    of a small bristle cleaning brush (MAJ_1888) which Olympus anticipates
    shipping no later than May 8, 2015”

    The key differences include the following:

    Pre cleaning

    . During immersion, raise and lower the elevator 3 times

    Manual cleaning

    . Additional brushing of the forceps elevator recess area using an
    additional brush (MAJ-1888 brush)

    . Additional flushing of forceps elevator recess area

    . Additional raising and lowering the forceps elevator

    Manual high level disinfection

    . Additional manual flushing steps and increased flushing volume of
    the endoscope channel and forceps recess area

    . Additional raising and lowering of the forceps elevator

    Regards

    Glenys

    Glenys Harrington

    Consultant

    Infection Control Consultancy (ICC)

    PO Box 5202

    Middle Park

    Victoria, 3206

    Australia

    M: +61 404 816 434

    infexion@ozemail.com.au

    ABN 47533508426

    Communication – New Reprocessing Instructions Validated

    MedWatch logo

    MedWatch – The FDA Safety Information and Adverse Event Reporting Program

    Duodenoscope Model TJF-Q180V by Olympus: FDA Safety Communication – New
    Reprocessing Instructions Validated

    Engineering, Patient

    for the TJF-Q180V duodenoscope to replace those provided in the original
    labeling. The FDA has reviewed these new reprocessing instructions and the
    validation data as part of its ongoing review of the 510(k), and recommends
    that any facilities that are using Olympus’ TJF-Q180V duodenoscope train
    staff on the new instructions and implement them as soon as possible.

    disinfect or sterilize reusable devices. The FDA is closely monitoring the
    possible association between reprocessed duodenoscopes and the transmission
    of infectious agents, including multidrug-resistant bacterial infections
    caused by Carbapenem-Resistant Enterobacteriaceae (CRE) such as Klebsiella
    species and Escherichia coli. If not properly reprocessed, residual body
    fluids and organic debris may remain in microscopic crevices of the device
    following an attempted cleaning and high level disinfection. If these
    residual fluids contain microbial contamination, subsequent patients may be
    exposed to serious infections. The FDA’s investigation into the possible
    association between inadequately reprocessed duodenoscopes and patient
    infections, including the agency’s recommendations for health care
    facilities, is more fully discussed in the following recent communications:
    February 2015 Safety Communication
    and FDA’s Updated Information for Healthcare Providers Regarding
    Duodenoscopes issued March 4, 2015.

    facilities and other users of the TJF-Q180V outlining the new, validated
    reprocessing instructions, and will soon be distributing revised user
    manuals. Key changes have been made to the Precleaning, Manual Cleaning,
    Manual High Level Disinfection reprocessing procedures for Olympus’
    TJF-Q180V duodenoscope. Please see the FDA Safety Communication
    for more details. In addition, FDA has the following recommendations
    for facilities and staff that use and reprocess the Olympus TJF-Q180V:

    *Implement the new manual cleaning and high level disinfection
    procedures for the Olympus TJF-Q180V duodenoscope in accordance with the
    manufacturer’s reprocessing instructions.
    *Train appropriate staff on Olympus’ new reprocessing instructions
    and implement them as soon as possible.
    *Contact Olympus directly with specific questions and concerns or to
    schedule a site visit with their Endoscopy Support Specialists Technical
    Assistance Center (TAC), 1-800-848-9024, option 1 Monday – Friday between
    7AM EST – 8 PM EST.

    Healthcare professionals and patients are encouraged to report adverse
    events or side effects related to the use of these products to the FDA’s
    MedWatch Safety Information and Adverse Event Reporting Program:

    *Complete and submit the report Online: http://www.fda.gov/MedWatch/report

    * Download form
    or call 1-800-332-1088 to request a reporting form, then complete and
    return to the address on the pre-addressed form, or submit by fax to
    1-800-FDA-0178

    Read the MedWatch Safety Alert, inclusing links to the FDA Safety
    Communication at:
    http://www.fda.gov/Safety/MedWatch/SafetyInformation/SafetyAlertsforHumanMed
    icalProducts/ucm440098.htm

    _____

    You are encouraged to report all serious adverse events and product quality
    problems to FDA MedWatch at http://www.fda.gov/medwatch/report.htm

    Follow MedWatch on Twitter
    twitter

    Update your subscriptions, modify your e-mail address, or stop subscriptions
    at any time on your Subscriber Preferences Page
    . You will need to use your email address
    to log in. If you have questions or problems with the subscription service,
    please visit subscriberhelp.govdelivery.com
    .

    This service is provided to you at no charge by U.S. Food
    & Drug Administration (FDA).

    FDA
    HHS

    _____

    This email was sent to infexion@ozemail.com.au using GovDelivery, on behalf
    of: U.S. Food & Drug Administration (FDA) . 10903 New Hampshire Ave . Silver
    Spring, MD 20993 . 800-439-1420

    Powered by GovDelivery

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    #72022
    Cath Murphy
    Participant

    Author:
    Cath Murphy

    Email:
    cath@INFECTIONCONTROLPLUS.COM.AU

    Organisation:

    State:

    Glenys et al

    Thank you for this really interesting commentary which again highlights the challenge of new technologies and our need for constant vigilance. As you are aware for many years Australia and I think parts of Canada have been outliers in terms of recommending regular sampling of scopes.

    APIC is currently writing an article for their Infection Preventionist magazine on duodenoscopes and have asked me to draft a few words on Australia’s approach to routine culturing as per GENCA Guidelines. APIC is particularly keen to share with their members the practical ways in which routine culturing is undertaken. I feel compelled to provide a broad overview and examples of different organisation’s practices and especially any difficulties w/ complying w/ sampling. I understand how 2 large SE QLD hospitals did it. I am familiar w/ the GENCA Guidelines and CHRISP protocols but I wondered if any on this list would be willing to offer insights. Dot points would be fine. I suspect that issues like:
    – time taken for results
    – lack of additional circulating scopes
    – access to micro labs
    – standardised procedure
    – action in the event of positive culture
    – clearance/ repeat reprocessing and culturing
    – ? need for Lookback, suspension of duodenoscopic services until clearanceetc
    may all be potentially problematic.

    Would list members who are willing to help me please post here ASAP as I am past the APIC deadline, or email me directly. Copies of local policy would be especially helpful .I am very happy to attribute anyone who provides insight (or have them remain anonymous as desired) and thank you in advance. I am also happy to share the final draft w/ any list members who may be interested.

    Regards
    Cath

    Dr Cathryn Murphy RN MPH PhD CIC
    Executive Director
    Infection Control Plus Pty Ltd
    Cath@infectioncontrolplus.com.au
    http://www.infectioncontrolplus.com.au
    [Description: twitter logo][Description: FB logo][Description: icp icon]

    Dear all,

    Find below the FDA notification of the New Reprocessing Instructions for the Olympus TJF-Q180V duodenoscopes.

    Attached is the Olympus letter to customers which details the new reprocessing instructions including;

    * supplementary flushing instructions

    * additional recess flushing and forceps elevator raising /lowering steps during pre-cleaning

    * manual cleaning

    * manual disinfection and endoscope rinsing and alcohol flushing.

    Olympus advise that “these new reprocessing procedures should be implemented as soon as possible” and that “the new cleaning procedure requires the use of a small bristle cleaning brush (MAJ_1888) which Olympus anticipates shipping no later than May 8, 2015”

    The key differences include the following:

    Pre cleaning

    * During immersion, raise and lower the elevator 3 times
    Manual cleaning

    * Additional brushing of the forceps elevator recess area using an additional brush (MAJ-1888 brush)

    * Additional flushing of forceps elevator recess area

    * Additional raising and lowering the forceps elevator
    Manual high level disinfection

    * Additional manual flushing steps and increased flushing volume of the endoscope channel and forceps recess area

    * Additional raising and lowering of the forceps elevator

    Regards

    Glenys

    Glenys Harrington
    Consultant
    Infection Control Consultancy (ICC)
    PO Box 5202
    Middle Park
    Victoria, 3206
    Australia
    M: +61 404 816 434
    infexion@ozemail.com.au
    ABN 47533508426

    [MedWatch logo][MedWatch – The FDA Safety Information and Adverse Event Reporting Program]

    Duodenoscope Model TJF-Q180V by Olympus: FDA Safety Communication – New Reprocessing Instructions Validated

    * Implement the new manual cleaning and high level disinfection procedures for the Olympus TJF-Q180V duodenoscope in accordance with the manufacturer’s reprocessing instructions.
    * Train appropriate staff on Olympus’ new reprocessing instructions and implement them as soon as possible.
    * Contact Olympus directly with specific questions and concerns or to schedule a site visit with their Endoscopy Support Specialists Technical Assistance Center (TAC), 1-800-848-9024, option 1 Monday – Friday between 7AM EST – 8 PM EST.

    Healthcare professionals and patients are encouraged to report adverse events or side effects related to the use of these products to the FDA’s MedWatch Safety Information and Adverse Event Reporting Program:

    * Complete and submit the report Online: http://www.fda.gov/MedWatch/report
    * Download form or call 1-800-332-1088 to request a reporting form, then complete and return to the address on the pre-addressed form, or submit by fax to 1-800-FDA-0178

    Read the MedWatch Safety Alert, inclusing links to the FDA Safety Communication at: http://www.fda.gov/Safety/MedWatch/SafetyInformation/SafetyAlertsforHumanMedicalProducts/ucm440098.htm

    ________________________________

    You are encouraged to report all serious adverse events and product quality problems to FDA MedWatch at http://www.fda.gov/medwatch/report.htm

    Follow MedWatch on Twitter

    Update your subscriptions, modify your e-mail address, or stop subscriptions at any time on your Subscriber Preferences Page. You will need to use your email address to log in. If you have questions or problems with the subscription service, please visit subscriberhelp.govdelivery.com.

    This service is provided to you at no charge by U.S. Food & Drug Administration (FDA).

    [FDA][HHS]

    ________________________________
    This email was sent to infexion@ozemail.com.au using GovDelivery, on behalf of: U.S. Food & Drug Administration (FDA) * 10903 New Hampshire Ave * Silver Spring, MD 20993 * 800-439-1420

    [Powered by GovDelivery]

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    #72182
    Glenys Harrington
    Participant

    Author:
    Glenys Harrington

    Email:
    infexion@ozemail.com.au

    Organisation:
    Infection Control Consultancy (ICC)

    State:

    Dear all,

    Further to my previous correspondence re: this topic, on 4th June 2015
    Olympus have released the attached URGENT Recall for Product Correction Re:
    TJF-Q180V Duodenoscope (TGA Ref #: RC-2015-RN-00475-1, ARTG Number: 210858)

    Please share with your colleagues working in Endoscopy Units

    Regards

    Glenys

    Glenys Harrington

    Consultant

    Infection Control Consultancy (ICC)

    PO Box 5202

    Middle Park

    Victoria, 3206

    Australia

    M: +61 404 816 434

    infexion@ozemail.com.au

    ABN 47533508426

    Of Glenys Harrington
    Endoscopic Retrograde Cholangiopancreatography – New Reprocessing
    Instructions for the Olympus TJF-Q180V duodenoscopes

    Dear all,

    Find below the FDA notification of the New Reprocessing Instructions for the
    Olympus TJF-Q180V duodenoscopes.

    Attached is the Olympus letter to customers which details the new
    reprocessing instructions including;

    . supplementary flushing instructions

    . additional recess flushing and forceps elevator raising /lowering
    steps during pre-cleaning

    . manual cleaning

    . manual disinfection and endoscope rinsing and alcohol flushing.

    Olympus advise that “these new reprocessing procedures should be implemented
    as soon as possible” and that “the new cleaning procedure requires the use
    of a small bristle cleaning brush (MAJ_1888) which Olympus anticipates
    shipping no later than May 8, 2015”

    The key differences include the following:

    Pre cleaning

    . During immersion, raise and lower the elevator 3 times

    Manual cleaning

    . Additional brushing of the forceps elevator recess area using an
    additional brush (MAJ-1888 brush)

    . Additional flushing of forceps elevator recess area

    . Additional raising and lowering the forceps elevator

    Manual high level disinfection

    . Additional manual flushing steps and increased flushing volume of
    the endoscope channel and forceps recess area

    . Additional raising and lowering of the forceps elevator

    Regards

    Glenys

    Glenys Harrington

    Consultant

    Infection Control Consultancy (ICC)

    PO Box 5202

    Middle Park

    Victoria, 3206

    Australia

    M: +61 404 816 434

    infexion@ozemail.com.au

    ABN 47533508426

    Communication – New Reprocessing Instructions Validated

    MedWatch logo

    MedWatch – The FDA Safety Information and Adverse Event Reporting Program

    Duodenoscope Model TJF-Q180V by Olympus: FDA Safety Communication – New
    Reprocessing Instructions Validated

    Engineering, Patient

    for the TJF-Q180V duodenoscope to replace those provided in the original
    labeling. The FDA has reviewed these new reprocessing instructions and the
    validation data as part of its ongoing review of the 510(k), and recommends
    that any facilities that are using Olympus’ TJF-Q180V duodenoscope train
    staff on the new instructions and implement them as soon as possible.

    disinfect or sterilize reusable devices. The FDA is closely monitoring the
    possible association between reprocessed duodenoscopes and the transmission
    of infectious agents, including multidrug-resistant bacterial infections
    caused by Carbapenem-Resistant Enterobacteriaceae (CRE) such as Klebsiella
    species and Escherichia coli. If not properly reprocessed, residual body
    fluids and organic debris may remain in microscopic crevices of the device
    following an attempted cleaning and high level disinfection. If these
    residual fluids contain microbial contamination, subsequent patients may be
    exposed to serious infections. The FDA’s investigation into the possible
    association between inadequately reprocessed duodenoscopes and patient
    infections, including the agency’s recommendations for health care
    facilities, is more fully discussed in the following recent communications:
    February 2015 Safety Communication
    and FDA’s Updated Information for Healthcare Providers Regarding
    Duodenoscopes issued March 4, 2015.

    facilities and other users of the TJF-Q180V outlining the new, validated
    reprocessing instructions, and will soon be distributing revised user
    manuals. Key changes have been made to the Precleaning, Manual Cleaning,
    Manual High Level Disinfection reprocessing procedures for Olympus’
    TJF-Q180V duodenoscope. Please see the FDA Safety Communication
    for more details. In addition, FDA has the following recommendations
    for facilities and staff that use and reprocess the Olympus TJF-Q180V:

    *Implement the new manual cleaning and high level disinfection
    procedures for the Olympus TJF-Q180V duodenoscope in accordance with the
    manufacturer’s reprocessing instructions.
    *Train appropriate staff on Olympus’ new reprocessing instructions
    and implement them as soon as possible.
    *Contact Olympus directly with specific questions and concerns or to
    schedule a site visit with their Endoscopy Support Specialists Technical
    Assistance Center (TAC), 1-800-848-9024, option 1 Monday – Friday between
    7AM EST – 8 PM EST.

    Healthcare professionals and patients are encouraged to report adverse
    events or side effects related to the use of these products to the FDA’s
    MedWatch Safety Information and Adverse Event Reporting Program:

    *Complete and submit the report Online: http://www.fda.gov/MedWatch/report

    * Download form
    or call 1-800-332-1088 to request a reporting form, then complete and
    return to the address on the pre-addressed form, or submit by fax to
    1-800-FDA-0178

    Read the MedWatch Safety Alert, inclusing links to the FDA Safety
    Communication at:
    http://www.fda.gov/Safety/MedWatch/SafetyInformation/SafetyAlertsforHumanMed
    icalProducts/ucm440098.htm

    _____

    You are encouraged to report all serious adverse events and product quality
    problems to FDA MedWatch at http://www.fda.gov/medwatch/report.htm

    Follow MedWatch on Twitter
    twitter

    Update your subscriptions, modify your e-mail address, or stop subscriptions
    at any time on your Subscriber Preferences Page
    . You will need to use your email address
    to log in. If you have questions or problems with the subscription service,
    please visit subscriberhelp.govdelivery.com
    .

    This service is provided to you at no charge by U.S. Food
    & Drug Administration (FDA).

    FDA
    HHS

    _____

    This email was sent to infexion@ozemail.com.au using GovDelivery, on behalf
    of: U.S. Food & Drug Administration (FDA) . 10903 New Hampshire Ave . Silver
    Spring, MD 20993 . 800-439-1420

    Powered by GovDelivery

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO
    NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is
    discouraged by ACIPC. If you wish to discuss specific reference to products
    or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives
    registration and login required.

    Replies to this message will be directed back to the list. To create a new
    message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to
    aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without
    the quotes) to listserv@aicalist.org.au

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

Viewing 3 posts - 1 through 3 (of 3 total)
  • The forum ‘Infexion Connexion’ is closed to new topics and replies.