    Dear All,

    Further to my previous postings on this topic:

    Australia

    . On 3/2/2016 Olympus Australia Pty Ltd and Olympus New Zealand Ltd
    in consultation with the TGA issued an “Urgent Recall for Product Correction
    for the TJF-Q180V duodenoscopes”

    o The company will perform correction to the forceps elevator mechanism.

    o The Operation and Reprocessing manuals have been updated.

    o Customers are expected to commence us of the updated TJF-Q180V Operation
    Manual and reprocessing Manual immediately.

    USA

    On February 19, 2016 the following FDA Safety Communication was released:

    . PENTAX Validates Reprocessing Instructions for ED-3490TK Video

    http://www.fda.gov/MedicalDevices/Safety/AlertsandNotices/ucm486772.htm

    Video Duodenoscope include a more rigorous protocol for pre-cleaning, manual
    cleaning, high-level disinfection and liquid chemical sterilization
    procedures. In addition, updated instructions include additional text,
    figures, cautions and warnings intended to clarify the validated
    reprocessing procedure. The Agency believes that when followed, these
    updated, validated reprocessing instructions demonstrate consistent and
    reliable cleaning, high-level disinfection and sterilization of PENTAX’s
    ED-3490TK’s Video Duodenoscope.

    http://www.fda.gov/Safety/MedWatch/SafetyInformation/SafetyAlertsforHumanMed
    icalProducts/ucm486887.htm

    On February 19, 2016 PENTAX sent letter to health care facilities and other
    users of the ED-3490TK Video Duodenoscope outlining the updated, validated
    reprocessing instructions – see attached

    I’m checking with the TGA to see if a letter with updated cleaning and
    disinfection instructions will be sent to Australian users of the PENTAX’s
    ED-3490TK’s Video Duodenoscope.

    Feel free to share this information with your clinical products advisors,
    colleagues in gastroenterology units, and risk managers.

    Regards

    Glenys

    Glenys Harrington

    Consultant

    Infection Control Consultancy (ICC)

    PO Box 5202

    Middle Park

    Victoria, 3206

    Australia

    M: +61 404 816 434

    infexion@ozemail.com.au

    ABN 47533508426

    Of Glenys Harrington
    Retrograde Cholangiopancreatography- Voluntary recall Olympus TJF-Q180V
    duodenoscopes

    Dear All,

    On 15/1/2016 the U.S. Food and Drug Administration cleared the Olympus
    TJF-Q180V duodenoscopes for modifications to the device design and labelling
    with the intention of reduce the risk of bacterial infections.

    http://www.fda.gov/NewsEvents/Newsroom/PressAnnouncements/ucm481956.htm

    Olympus America Inc (“OAI”) has posted a voluntary recall to replace the
    forceps elevator mechanism in all TJF-Q180V duodenoscopes (a flexible
    gastrointestinal endoscope used in procedures such as endoscopic retrograde
    cholangiopancreatography (ERCP) – see attached letter

    Staff in you gastroenterology units will be interested in this recall and
    should contact the manufacturer for additional information.

    Regards

    Glenys

    Glenys Harrington

    Consultant

    Infection Control Consultancy (ICC)

    PO Box 5202

    Middle Park

    Victoria, 3206

    Australia

    M: +61 404 816 434

    infexion@ozemail.com.au

    ABN 47533508426

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO
    NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is
    discouraged by ACIPC. If you wish to discuss specific reference to products
    or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives
    registration and login required.

    Replies to this message will be directed back to the list. To create a new
    message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to
    aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without
    the quotes) to listserv@aicalist.org.au

    MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

    The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

Viewing 1 post (of 1 total)
  • The forum ‘Infexion Connexion’ is closed to new topics and replies.