Home › Forums › Infexion Connexion › Does use of plastic tray for instrument reprocessing comply with AS/NZS 4187 standard?
- This topic has 0 replies, 3 voices, and was last updated 4 years, 2 months ago by Liberali, Sharon (Health).
-
AuthorPosts
-
22/09/2020 at 2:48 pm #77324Kugathas, Kavitha (Health)Participant
Author:
Kugathas, Kavitha (Health)Email:
Kavitha.Kugathas@ACT.GOV.AUOrganisation:
State:
OFFICIAL
Hi Team,
We use plastic perforated trays for instrument (RMD) reprocessing. Does this comply with AS/NZS 4187 standard? The AS/NZS 4187 standard is vague. We have been recommended to replace all plastic perforated trays with open wire mesh tray. This will be a very expensive option for us.
We have validated the plastic perforated trays in ultrasonic, washer disinfectors and sterilisers during the performance qualification also each year during re- performance qualification.
Regards, Kavitha
Kavitha Kugathas | Director
Food and Sterilising Services |Sterilising Services
Canberra Health Services | ACT Government[Canberra Health Services_RGB]
RELIABLE | PROGRESSIVE | RESPECTFUL | KIND
———————————————————————–
This email, and any attachments, may be confidential and also privileged. If you are not the intended recipient, please notify the sender and delete all copies of this transmission along with any attachments immediately. You should not copy or use it for any purpose, nor disclose its contents to any other person.
———————————————————————–MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.
The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.
Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.
Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au
To send a message to the list administrator send an email to admin@acipc.org.au
You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au
22/09/2020 at 3:05 pm #77325AnonymousInactiveAuthor:
AnonymousOrganisation:
State:
Hi Kavitha,
AS/NZS4187 does not dictate the types of containment systems [trays] that you use for your RMDs.
Plastic trays have been in wide spread use for decades in Australia, however as those of us working in reprocessing know, the solid sides of the trays and the thickness of the squares in the bottom of the trays can easily impede cleaning effectiveness.
Similarly, the plastic trays are not ideal for ensuring that the load in a washer-disinfector is easily dried and we can experience problems with wet loads in the steriliser for the same reasons.
In short, using metal mesh trays for your RMDs is considered best practice and I would encourage you to work towards eliminating the plastic trays over time and as budget allows.
Kind Regards
Terry McAuley
Director
MSc Medical Device DecontaminationPO BOX 2249, Greenvale, VIC Australia 3059
[cid:image001.png@01D690F1.C2DC1690]
I endeavour to achieve a sensible work-life balance: There is no need to reply to this email from you outside of your normal working hours. Please expect the same from me.
CONFIDENTIAL COMMUNICATION: The information contained in this message may contain confidential information intended only for the use of the individual or entity named above. If the reader of this message is not the intended recipient, you are hereby notified that any dissemination, distribution or duplication of this transmission is strictly prohibited. If you have received this communication in error, please notify us by telephone or email immediately and return the original message to us or destroy all printed and electronic copies. Nothing in this transmission constitutes an agreement of any kind unless otherwise expressly indicated.
OFFICIAL
Hi Team,
We use plastic perforated trays for instrument (RMD) reprocessing. Does this comply with AS/NZS 4187 standard? The AS/NZS 4187 standard is vague. We have been recommended to replace all plastic perforated trays with open wire mesh tray. This will be a very expensive option for us.
We have validated the plastic perforated trays in ultrasonic, washer disinfectors and sterilisers during the performance qualification also each year during re- performance qualification.
Regards, Kavitha
Kavitha Kugathas | Director
Food and Sterilising Services |Sterilising Services
Canberra Health Services | ACT Government[Canberra Health Services_RGB]
RELIABLE | PROGRESSIVE | RESPECTFUL | KIND
———————————————————————–
This email, and any attachments, may be confidential and also privileged. If you are not the intended recipient, please notify the sender and delete all copies of this transmission along with any attachments immediately. You should not copy or use it for any purpose, nor disclose its contents to any other person.
———————————————————————–
MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.
Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.
Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au
To send a message to the list administrator send an email to admin@acipc.org.au
You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au
MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.
The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.
Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.
Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au
To send a message to the list administrator send an email to admin@acipc.org.au
You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au
22/09/2020 at 3:16 pm #77326Liberali, Sharon (Health)ParticipantAuthor:
Liberali, Sharon (Health)Email:
Sharon.Liberali@SA.GOV.AUOrganisation:
State:
Hi Kavitha,
I assume that you when you mention the ‘plastic perforated trays’ you are referring to the packaging system, which sits under Section5.5 as well as Guidance A5.5 within AS/NZS 4187.
This states that the packaging system needs to conform to ISO 11607-1 and ISO 11607-2 and that it needs to be compatible with the sterilizing process. It needs to allow the removal of air, ingress and egress of the sterilizing agent and removal of water vapour.
My background is in dentistry, and we commonly use ‘plastic packaging systems’ e.g. cassettes for reprocessing of our semi-critical instruments. We also use stainless steel packaging systems, and disposable packaging systems (steri bags). In addition, like you all of these different packaging systems are validated and re-validated as part of PQ.
I sit on the HE-023 Standards committee which writes 4187, and to my knowledge there is nothing in the standard which necessitates the need for a specific type of packaging system.
Sharon
A/Prof. Sharon Liberali
BDS(Adel); Grad.Dip.Clin.Dent; MScMed (OFP); D.Clin.Dent (SND);FRACDS (SND); FICD; FPFA.Specialist, Special Needs Dentistry
Director, Special Needs Unit, Adelaide Dental Hospital, SA Dental Service.
Program Convenor, Special Needs Dentistry, School of Dentistry, University of Adelaide.The information in this e-mail may be confidential and/or legally privileged. It is intended solely for the addressee.
Access to this e-mail by anyone else is unauthorised. If you are not the intended recipient, any disclosure, copying, distribution or any action taken or omitted to be taken in reliance on it, is prohibited and may be unlawful.OFFICIAL
Hi Team,
We use plastic perforated trays for instrument (RMD) reprocessing. Does this comply with AS/NZS 4187 standard? The AS/NZS 4187 standard is vague. We have been recommended to replace all plastic perforated trays with open wire mesh tray. This will be a very expensive option for us.
We have validated the plastic perforated trays in ultrasonic, washer disinfectors and sterilisers during the performance qualification also each year during re- performance qualification.
Regards, Kavitha
Kavitha Kugathas | Director
Food and Sterilising Services |Sterilising Services
Canberra Health Services | ACT Government[Canberra Health Services_RGB]
RELIABLE | PROGRESSIVE | RESPECTFUL | KIND
———————————————————————–
This email, and any attachments, may be confidential and also privileged. If you are not the intended recipient, please notify the sender and delete all copies of this transmission along with any attachments immediately. You should not copy or use it for any purpose, nor disclose its contents to any other person.
———————————————————————–
MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.
Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.
Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au
To send a message to the list administrator send an email to admin@acipc.org.au
You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au
MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.
The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.
Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.
Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au
To send a message to the list administrator send an email to admin@acipc.org.au
You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au
-
AuthorPosts
- The forum ‘Infexion Connexion’ is closed to new topics and replies.