CT Contrast Injecting System

Home Forums Infexion Connexion CT Contrast Injecting System

 | Click to Receive Email Notifications of Posts
  • This topic has 1 reply, 1 voice, and was last updated 12 years ago by Anonymous.
Viewing 5 posts - 1 through 5 (of 5 total)
  • Author
    Posts
  • #69165
    Fiona de Sousa
    Participant

    Author:
    Fiona de Sousa

    Position:

    Organisation:

    State:

    Hi All,

    I have been asked to evaluate a new system for injecting CT contrast to
    determine its suitability for our facility. The transflux system allows
    for multi use of syringes, common reusable tubing (both changed daily)
    and a connector tubing that is changed for every patient. It has TGA
    approval. It has been used for a number of years in Europe.

    I am concerned about the potential for line contamination from blood as
    well as the sterility of the connector ends. I would like to hear from
    anyone who has previously evaluated this system or has used it.

    Kind Regards,

    Fiona De Sousa

    Infection Prevention & Control Coordinator

    Sydney Adventist Hospital

    Fiona.Desousa@sah.org.au

    185 Fox Valley Road, Wahroonga, NSW, 2076

    If you are not the intended recipient you are hereby notified that any dissemination, distribution or reproduction of this message
    is prohibited. If you have received this message in error please notify the sender immediately, then destroy the original message.
    Any views expressed in this message are solely those of the individual sender, except where the sender is specifically authorised
    by Sydney Adventist Hospital to state that they are the views of Sydney Adventist Hospital.
    _____________________________________________________________________
    This e-mail has been scanned for viruses by Symantec Hosted Services
    Scanning Services – powered by MessageLabs. For further information
    visit http://www.messagelabs.com

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    #69168
    Radion Ingman
    Participant

    Author:
    Radion Ingman

    Position:

    Organisation:

    State:

    Hi Fiona,

    You can use IV, 0.2 micron filter, between common reusable tubing and
    connector tubing that is connected to a patient. With that both the
    infusion system and the patient will be protected from crossover
    contamination. I can recommend you a number of IV filters that are already
    successfully used for many procedures, including CT contrast infusions.
    Please do not hesitate to contact me at any time convenient for further
    information.

    Warm Regards,

    Radion Ingman
    Territory Manager NSW/ACT
    Pall Australia Pty Ltd
    Suite 1 & 2, 36 Fitzroy St
    Surry Hills NSW 2010
    http://www.pall.com

    Fiona de Sousa
    Sent by: ACIPC Infexion Connexion
    10/07/2012 05:41 PM
    Please respond to
    ACIPC Infexion Connexion

    To
    AICALIST@AICALIST.ORG.AU
    cc

    Subject
    CT Contrast Injecting System

    Hi All,

    I have been asked to evaluate a new system for injecting CT contrast to
    determine its suitability for our facility. The transflux system allows
    for multi use of syringes, common reusable tubing (both changed daily) and
    a connector tubing that is changed for every patient. It has TGA
    approval. It has been used for a number of years in Europe.

    I am concerned about the potential for line contamination from blood as
    well as the sterility of the connector ends. I would like to hear from
    anyone who has previously evaluated this system or has used it.

    Kind Regards,

    Fiona De Sousa
    Infection Prevention & Control Coordinator
    Sydney Adventist Hospital
    Fiona.Desousa@sah.org.au
    185 Fox Valley Road, Wahroonga, NSW, 2076

    information intended for the addressee named above.
    If you are not the intended recipient you are hereby notified that any
    dissemination, distribution or reproduction of this message
    is prohibited. If you have received this message in error please notify
    the sender immediately, then destroy the original message.
    Any views expressed in this message are solely those of the individual
    sender, except where the sender is specifically authorised
    by Sydney Adventist Hospital to state that they are the views of Sydney
    Adventist Hospital.
    _____________________________________________________________________
    This e-mail has been scanned for viruses by Symantec Hosted Services
    Scanning Services – powered by MessageLabs. For further information
    visit http://www.messagelabs.com
    Messages posted to this list are solely the opinion of the authors, and do
    not represent the opinion of ACIPC.
    Archive of all messages are available at http://aicalist.org.au/archives
    registration and login required.
    Replies to this message will be directed back to the list. To create a new
    message send an email to aicalist@aicalist.org.au
    To send a message to the list administrator send an email to
    aicalist-request@aicalist.org.au.
    You can unsubscribe from this list be sending ‘signoff aicalist’ (without
    the quotes) to listserv@aicalist.org.au

    ——————————————————————————–

    Attention:

    This communication may contain information that is confidential,
    privileged and/or exempt from disclosure under applicable law.
    If you are not the intended recipient, please notify the sender
    immediately and delete the original, all attachments, and all
    copies of this communication.

    ——————————————————————————–

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    #69169
    Sheila Hamm
    Participant

    Author:
    Sheila Hamm

    Position:

    Organisation:

    State:

    Hi Fiona,

    I looked at a very similar system (if not the same one) a couple of
    years ago and I had exactly the same concerns as you.

    We elected to stay with the existing system, it’s entirely single pt
    use, and staff were happy with it, so there were no real benefits, cost
    or otherwise to changing.

    Regards

    Sheila

    Sheila Hamm

    Clinical Nurse Specialist Infection Control | Nepean Hospital
    PO Box 63 Penrith NSW 2751
    Pager 17254 l Tel 02 4734 2228 | Fax 02 4734 2517 |
    sheila.hamm@swahs.health.nsw.gov.au

    http://www.health.nsw.gov.au

    Behalf Of Fiona de Sousa

    Hi All,

    I have been asked to evaluate a new system for injecting CT contrast to
    determine its suitability for our facility. The transflux system allows
    for multi use of syringes, common reusable tubing (both changed daily)
    and a connector tubing that is changed for every patient. It has TGA
    approval. It has been used for a number of years in Europe.

    I am concerned about the potential for line contamination from blood as
    well as the sterility of the connector ends. I would like to hear from
    anyone who has previously evaluated this system or has used it.

    Kind Regards,

    Fiona De Sousa

    Infection Prevention & Control Coordinator

    Sydney Adventist Hospital

    Fiona.Desousa@sah.org.au

    185 Fox Valley Road, Wahroonga, NSW, 2076

    information intended for the addressee named above.
    If you are not the intended recipient you are hereby notified that any
    dissemination, distribution or reproduction of this message
    is prohibited. If you have received this message in error please notify
    the sender immediately, then destroy the original message.
    Any views expressed in this message are solely those of the individual
    sender, except where the sender is specifically authorised
    by Sydney Adventist Hospital to state that they are the views of Sydney
    Adventist Hospital.
    _____________________________________________________________________
    This e-mail has been scanned for viruses by Symantec Hosted Services
    Scanning Services – powered by MessageLabs. For further information
    visit http://www.messagelabs.com

    ___________________________________

    This email was scanned for viruses on 07/10/12 at 17:41:00

    If you have any questions, please call Statewide Servicedesk on 1800 285
    533 and ask for a REQUEST to be logged for review by the ITS
    Telecommunications Team.

    Messages posted to this list are solely the opinion of the authors, and
    do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives
    – registration and login required.

    Replies to this message will be directed back to the list. To create a
    new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to
    aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’
    (without the quotes) to listserv@aicalist.org.au

    ___________________________________

    Unless you are the intended recipient any unauthorised use, dissemination,further distribution or reproduction of this communication in any form whatsoever, is strictly prohibited.

    If this communication has been sent to you in error, please notify the sender by return e-mail and delete and/or destroy your copy of this communication (including attachments).

    Any views expressed in this communication are those of the individual sender, except where the sender states them to be the views of the Nepean Blue Mountains Local Health District/Western Sydney Local Health District.

    Unless otherwise expressed, it is not represented, warranted or guaranteed that the integrity of this communication has been maintained nor that the communication is free of virus, errors or interference.

    07/11/12 – 15:14:57

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    #69170
    Anonymous
    Inactive

    Author:
    Anonymous

    Position:

    Organisation:

    State:

    Hi Fiona,
    We have looked a a number of automated contrast delivery systems, and had the same concern about multi-patient consumables. The Acist system was one we evaluated recently, this system has a one way/anti-reflux valve in the single patient consumable tubing, which was not the case when we evaluated a similar pump about two years ago. With the addition of the one way valve and information that similar pumps are in use in other Brisbane hospitals without any increase in infection, our infection control team agreed to trial.
    The major reason for changing to the automated system was a number of nursing staff injuries due to RSI from repeated injection of contrast by hand. The other benefit is that the patient receives less contrast using the automated system, which benefits the patient and also has a cost benefit to the organisation. Hope this information is useful to you.

    Regards
    Kathy

    Kathy Taylor CICP
    Infection Control Manager
    The Wesley Hospital
    PO Box 499,
    Toowong, Qld 4066
    07 3232 7558
    katherine.taylor@uchealth.com.au

    ________________________________

    Hi All,

    I have been asked to evaluate a new system for injecting CT contrast to determine its suitability for our facility. The transflux system allows for multi use of syringes, common reusable tubing (both changed daily) and a connector tubing that is changed for every patient. It has TGA approval. It has been used for a number of years in Europe.

    I am concerned about the potential for line contamination from blood as well as the sterility of the connector ends. I would like to hear from anyone who has previously evaluated this system or has used it.

    Kind Regards,

    Fiona De Sousa
    Infection Prevention & Control Coordinator
    Sydney Adventist Hospital
    Fiona.Desousa@sah.org.au
    185 Fox Valley Road, Wahroonga, NSW, 2076

    If you are not the intended recipient you are hereby notified that any dissemination, distribution or reproduction of this message
    is prohibited. If you have received this message in error please notify the sender immediately, then destroy the original message.
    Any views expressed in this message are solely those of the individual sender, except where the sender is specifically authorised
    by Sydney Adventist Hospital to state that they are the views of Sydney Adventist Hospital.
    _____________________________________________________________________
    This e-mail has been scanned for viruses by Symantec Hosted Services
    Scanning Services – powered by MessageLabs. For further information
    visit http://www.messagelabs.com

    *********************************************************************************

    Although this email has been scanned for viruses please ensure that due care

    is taken before opening any attachments. Do you know the sender?

    Do not open any attachments that you are not expecting!

    *********************************************************************************

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    _________________________________________________________________
    Uniting Care Health Email Disclaimer: http://www.uchealth.com.au/disclaimer

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    #69171
    Fiona de Sousa
    Participant

    Author:
    Fiona de Sousa

    Position:

    Organisation:

    State:

    Thanks Kathy,

    How do you ensure that the line connector does not become contaminated
    with the multiple access that is required for the reusable part of the
    system?

    Kind regards,

    Fiona De Sousa

    Infection Prevention & Control Coordinator

    Sydney Adventist Hospital

    Fiona.Desousa@sah.org.au

    185 Fox Valley Road, Wahroonga, NSW, 2076

    ________________________________

    Behalf Of Katherine Taylor

    Hi Fiona,

    We have looked a a number of automated contrast delivery systems, and
    had the same concern about multi-patient consumables. The Acist system
    was one we evaluated recently, this system has a one way/anti-reflux
    valve in the single patient consumable tubing, which was not the case
    when we evaluated a similar pump about two years ago. With the addition
    of the one way valve and information that similar pumps are in use in
    other Brisbane hospitals without any increase in infection, our
    infection control team agreed to trial.

    The major reason for changing to the automated system was a number of
    nursing staff injuries due to RSI from repeated injection of contrast by
    hand. The other benefit is that the patient receives less contrast
    using the automated system, which benefits the patient and also has a
    cost benefit to the organisation. Hope this information is useful to
    you.

    Regards

    Kathy

    Kathy Taylor CICP

    Infection Control Manager

    The Wesley Hospital

    PO Box 499,

    Toowong, Qld 4066

    07 3232 7558

    katherine.taylor@uchealth.com.au

    ________________________________

    Behalf Of Fiona de Sousa

    Hi All,

    I have been asked to evaluate a new system for injecting CT contrast to
    determine its suitability for our facility. The transflux system allows
    for multi use of syringes, common reusable tubing (both changed daily)
    and a connector tubing that is changed for every patient. It has TGA
    approval. It has been used for a number of years in Europe.

    I am concerned about the potential for line contamination from blood as
    well as the sterility of the connector ends. I would like to hear from
    anyone who has previously evaluated this system or has used it.

    Kind Regards,

    Fiona De Sousa

    Infection Prevention & Control Coordinator

    Sydney Adventist Hospital

    Fiona.Desousa@sah.org.au

    185 Fox Valley Road, Wahroonga, NSW, 2076

    information intended for the addressee named above.
    If you are not the intended recipient you are hereby notified that any
    dissemination, distribution or reproduction of this message
    is prohibited. If you have received this message in error please notify
    the sender immediately, then destroy the original message.
    Any views expressed in this message are solely those of the individual
    sender, except where the sender is specifically authorised
    by Sydney Adventist Hospital to state that they are the views of Sydney
    Adventist Hospital.
    _____________________________________________________________________
    This e-mail has been scanned for viruses by Symantec Hosted Services
    Scanning Services – powered by MessageLabs. For further information
    visit http://www.messagelabs.com

    ************************************************************************
    *********

    Although this email has been scanned for viruses please ensure that due
    care

    is taken before opening any attachments. Do you know the sender?

    Do not open any attachments that you are not expecting!

    ************************************************************************
    *********

    Messages posted to this list are solely the opinion of the authors, and
    do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives
    – registration and login required.

    Replies to this message will be directed back to the list. To create a
    new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to
    aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’
    (without the quotes) to listserv@aicalist.org.au

    _________________________________________________________________

    Uniting Care Health Email Disclaimer:
    http://www.uchealth.com.au/disclaimer

    Messages posted to this list are solely the opinion of the authors, and
    do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives
    – registration and login required.

    Replies to this message will be directed back to the list. To create a
    new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to
    aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’
    (without the quotes) to listserv@aicalist.org.au

    If you are not the intended recipient you are hereby notified that any dissemination, distribution or reproduction of this message
    is prohibited. If you have received this message in error please notify the sender immediately, then destroy the original message.
    Any views expressed in this message are solely those of the individual sender, except where the sender is specifically authorised
    by Sydney Adventist Hospital to state that they are the views of Sydney Adventist Hospital.
    _____________________________________________________________________
    This e-mail has been scanned for viruses by Symantec Hosted Services
    Scanning Services – powered by MessageLabs. For further information
    visit http://www.messagelabs.com

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

Viewing 5 posts - 1 through 5 (of 5 total)
  • The forum ‘Infexion Connexion’ is closed to new topics and replies.