Benchtop steriliser water and steam quality specifications.

Home Forums Infexion Connexion Benchtop steriliser water and steam quality specifications.

 | Click to Receive Email Notifications of Posts
Viewing 5 posts - 1 through 5 (of 5 total)
  • Author
    Posts
  • #69952
    Beth Bint
    Participant

    Author:
    Beth Bint

    Position:

    Organisation:

    State:

    Hi All

    Hoping someone can offer some advice.

    We have not be able to find definitive specifications for water and steam quality required for benchtop sterilisers (Dental Autoclaves).

    We have two questions:

    1. Some manufacturers suggest the use of distilled or de-ionised water. If a water distiller is used how is the water quality controlled when refilling the reservoir?

    2. What are the steam quality specifications for benchtop sterilisers, and where is the reference for these?

    Thank you for any assistance you can provide.

    Beth

    Beth Bint

    Clinical Nurse Consultant | Infection Management and Control Service
    Level 1 Lawson House, Wollongong Hospital 2500, NSW
    Tel. 02 4222 5869 | Fax. 02 4222 5367 | beth.bint@sesiahs.health.nsw.gov.au

    [cid:image001.jpg@01CE40F2.33F57270]

    ———————————————————————————————

    Illawarra Shoalhaven Local Health District, South East Sydney Local Health District and Sydney Children’s Hospital Network (Randwick Campus) Confidentiality Notice

    This email, and the files transmitted with it, are confidential and intended solely for the use of the individual or entity to whom they are addressed. If you are not the intended recipient, you are not permitted to distribute or use this email or any of its attachments in any way. We also request that you advise the sender of the incorrect addressing.

    This email message has been virus-scanned. Although no computer viruses were detected, Illawarra Shoalhaven Local Health District, South East Sydney Local Health District and Sydney Children’s Hospital Network (Randwick Campus) accept no liability for any consequential damage resulting from email containing any computer viruses.

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    #69959
    McEwan, Melissa MS
    Participant

    Author:
    McEwan, Melissa MS

    Position:

    Organisation:

    State:

    UNCLASSIFIED

    Hi Beth
    I am not an expert in this area but I know that the staff use distilled
    water in the Lisa sterilisers that are used in dental .
    The steam quality specifications – isn’t this your Bowie Dick and Helix
    test.
    The water reservoir in the Lisa is cleaned following the manufacturer
    instructions ( I believe this is an alcohol solution)
    If you need any further help with this, I can put you in touch with the
    dental manager
    Regards

    Melissa McEwan RN, BN, Grad Cert Infect Control
    Quality Manager
    Contractor to Defence
    Wagga Wagga
    02 69338338
    Private mobile 0428 753783
    melissa.mcewan@defence.gov.au

    and is subject to the jurisdiction of section 70 of the Crimes Act 1914.
    If you have received this email in error, you are requested to contact
    the sender and delete the email.

    ________________________________

    Behalf Of Beth Bint

    Hi All

    Hoping someone can offer some advice.

    We have not be able to find definitive specifications for water and
    steam quality required for benchtop sterilisers (Dental Autoclaves).

    We have two questions:

    1. Some manufacturers suggest the use of distilled or de-ionised
    water. If a water distiller is used how is the water quality controlled
    when refilling the reservoir?

    2. What are the steam quality specifications for benchtop
    sterilisers, and where is the reference for these?

    Thank you for any assistance you can provide.

    Beth

    Beth Bint

    Clinical Nurse Consultant | Infection Management and Control Service
    Level 1 Lawson House, Wollongong Hospital 2500, NSW
    Tel. 02 4222 5869 | Fax. 02 4222 5367 |
    beth.bint@sesiahs.health.nsw.gov.au

    http://www.health.nsw.gov.au/images/communications/e-signatures/images/N
    SW-Health-Illawarra-Shoalhaven-LHD.jpg

    ————————————————————————
    ———————

    Illawarra Shoalhaven Local Health District, South East Sydney Local
    Health District and Sydney Children’s Hospital Network (Randwick Campus)
    Confidentiality Notice

    This email, and the files transmitted with it, are confidential and
    intended solely for the use of the individual or entity to whom they are
    addressed. If you are not the intended recipient, you are not permitted
    to distribute or use this email or any of its attachments in any way. We
    also request that you advise the sender of the incorrect addressing.

    This email message has been virus-scanned. Although no computer viruses
    were detected, Illawarra Shoalhaven Local Health District, South East
    Sydney Local Health District and Sydney Children’s Hospital Network
    (Randwick Campus) accept no liability for any consequential damage
    resulting from email containing any computer viruses.
    Messages posted to this list are solely the opinion of the authors, and
    do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives
    – registration and login required.

    Replies to this message will be directed back to the list. To create a
    new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to
    aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’
    (without the quotes) to listserv@aicalist.org.au

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    #69961
    Anonymous
    Inactive

    Author:
    Anonymous

    Position:

    Organisation:

    State:

    Hi Beth,

    Distilled or deionised water is used in the reservoir to limit the
    contaminants in the steam generated – particularly in terms of the chemical
    purity, however the microbiological purity may still be a concern,
    particularly in situations where water sits for prolonged periods of time.

    I have issues with electrically operated water distillers as they are rarely
    cleaned nor is the microbiological purity or endotoxin levels ever
    monitored. I prefer people to buy in distilled water, as there should be
    quality control systems in place from the manufacturer of this product.

    Steam quality is not typically monitored on small steam sterilisers. This
    would be very difficult to do.

    The Bowie-Dick type test is an air removal and steam penetration test. Some
    products indicate that the result can be used to indicate steam quality –
    i.e. wet or superheated steam conditions, however this is not the primary
    function of the test.

    Regards

    Terry McAuley

    Sterilisation & Infection Control Consultant

    STEAM Consulting

    E: terry@steamconsulting.com.au

    W: http://www.steamconsulting.com.au

    A: PO BOX 779

    Endeavour Hills

    VIC Australia 3802

    CONFIDENTIAL COMMUNICATION: The information contained in this message may
    contain confidential information intended only for the use of the individual
    or entity named above. If the reader of this message is not the intended
    recipient, you are hereby notified that any dissemination, distribution or
    duplication of this transmission is strictly prohibited. If you have
    received this communication in error, please notify us by telephone or email
    immediately and return the original message to us or destroy all printed and
    electronic copies. Nothing in this transmission constitutes an agreement of
    any kind unless otherwise expressly indicated.

    Of McEwan, Melissa MS
    [SEC=UNCLASSIFIED]

    UNCLASSIFIED

    Hi Beth

    I am not an expert in this area but I know that the staff use distilled
    water in the Lisa sterilisers that are used in dental .

    The steam quality specifications – isn’t this your Bowie Dick and Helix
    test.

    The water reservoir in the Lisa is cleaned following the manufacturer
    instructions ( I believe this is an alcohol solution)

    If you need any further help with this, I can put you in touch with the
    dental manager

    Regards

    Melissa McEwan RN, BN, Grad Cert Infect Control

    Quality Manager

    Contractor to Defence

    Wagga Wagga

    02 69338338

    Private mobile 0428 753783

    melissa.mcewan@defence.gov.au

    is subject to the jurisdiction of section 70 of the Crimes Act 1914. If you
    have received this email in error, you are requested to contact the sender
    and delete the email.

    _____

    Of Beth Bint

    Hi All

    Hoping someone can offer some advice.

    We have not be able to find definitive specifications for water and steam
    quality required for benchtop sterilisers (Dental Autoclaves).

    We have two questions:

    1. Some manufacturers suggest the use of distilled or de-ionised
    water. If a water distiller is used how is the water quality controlled
    when refilling the reservoir?

    2. What are the steam quality specifications for benchtop sterilisers,
    and where is the reference for these?

    Thank you for any assistance you can provide.

    Beth

    Beth Bint

    Clinical Nurse Consultant | Infection Management and Control Service
    Level 1 Lawson House, Wollongong Hospital 2500, NSW
    Tel. 02 4222 5869 | Fax. 02 4222 5367 | beth.bint@sesiahs.health.nsw.gov.au

    http://www.health.nsw.gov.au/images/communications/e-signatures/images/NSW-H
    ealth-Illawarra-Shoalhaven-LHD.jpg

    —————————————————————————-
    —————–

    Illawarra Shoalhaven Local Health District, South East Sydney Local Health
    District and Sydney Children’s Hospital Network (Randwick Campus)
    Confidentiality Notice

    This email, and the files transmitted with it, are confidential and intended
    solely for the use of the individual or entity to whom they are addressed.
    If you are not the intended recipient, you are not permitted to distribute
    or use this email or any of its attachments in any way. We also request that
    you advise the sender of the incorrect addressing.

    This email message has been virus-scanned. Although no computer viruses were
    detected, Illawarra Shoalhaven Local Health District, South East Sydney
    Local Health District and Sydney Children’s Hospital Network (Randwick
    Campus) accept no liability for any consequential damage resulting from
    email containing any computer viruses.

    Messages posted to this list are solely the opinion of the authors, and do
    not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives
    registration and login required.

    Replies to this message will be directed back to the list. To create a new
    message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to
    aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without
    the quotes) to listserv@aicalist.org.au

    Messages posted to this list are solely the opinion of the authors, and do
    not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives
    registration and login required.

    Replies to this message will be directed back to the list. To create a new
    message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to
    aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without
    the quotes) to listserv@aicalist.org.au

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    #69962
    Roel Castillo
    Participant

    Author:
    Roel Castillo

    Position:

    Organisation:

    State:

    Hi Beth

    Please refer to AS 4815 for office based reprocessing and AS 4187 for healthcare facilities reprocessing of reusable medical devices. Bowie Dick test and Helix test are process challenges to ascertain your sterilisers function as required, depending on the steriliser, BD is used to measure air removal and steam penetration, helix challenges the efficacy of the method in attaining the set parameters for sterilisation.
    Steam quality as required should have 97% dryness fraction and water quality in today’s standards would recommend RO quality, but in the current standards it does not require it yet. A good perusal of the aforementioned standards is worthwhile.

    Cheers

    Roel Castillo
    Project Officer – SSD

    UNCLASSIFIED
    Hi Beth

    I am not an expert in this area but I know that the staff use distilled water in the Lisa sterilisers that are used in dental .

    The steam quality specifications – isn’t this your Bowie Dick and Helix test.

    The water reservoir in the Lisa is cleaned following the manufacturer instructions ( I believe this is an alcohol solution)

    If you need any further help with this, I can put you in touch with the dental manager

    Regards

    Melissa McEwan RN, BN, Grad Cert Infect Control
    Quality Manager
    Contractor to Defence
    Wagga Wagga
    02 69338338
    Private mobile 0428 753783
    melissa.mcewan@defence.gov.au

    ________________________________
    Hi All

    Hoping someone can offer some advice.

    We have not be able to find definitive specifications for water and steam quality required for benchtop sterilisers (Dental Autoclaves).

    We have two questions:

    1. Some manufacturers suggest the use of distilled or de-ionised water. If a water distiller is used how is the water quality controlled when refilling the reservoir?

    2. What are the steam quality specifications for benchtop sterilisers, and where is the reference for these?

    Thank you for any assistance you can provide.

    Beth

    Beth Bint

    Clinical Nurse Consultant | Infection Management and Control Service
    Level 1 Lawson House, Wollongong Hospital 2500, NSW
    Tel. 02 4222 5869 | Fax. 02 4222 5367 | beth.bint@sesiahs.health.nsw.gov.au

    [http://www.health.nsw.gov.au/images/communications/e-signatures/images/NSW-Health-Illawarra-Shoalhaven-LHD.jpg]

    ———————————————————————————————

    Illawarra Shoalhaven Local Health District, South East Sydney Local Health District and Sydney Children’s Hospital Network (Randwick Campus) Confidentiality Notice

    This email, and the files transmitted with it, are confidential and intended solely for the use of the individual or entity to whom they are addressed. If you are not the intended recipient, you are not permitted to distribute or use this email or any of its attachments in any way. We also request that you advise the sender of the incorrect addressing.

    This email message has been virus-scanned. Although no computer viruses were detected, Illawarra Shoalhaven Local Health District, South East Sydney Local Health District and Sydney Children’s Hospital Network (Randwick Campus) accept no liability for any consequential damage resulting from email containing any computer viruses.

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    ______________________________________________________________________
    This email has been scanned by the Symantec Email Security.cloud service.
    For more information please visit http://www.symanteccloud.com
    ______________________________________________________________________
    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    ______________________________________________________________________
    This email has been scanned by the Symantec Email Security.cloud service.
    For more information please visit http://www.symanteccloud.com
    ______________________________________________________________________

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    #69967
    Beth Bint
    Participant

    Author:
    Beth Bint

    Position:

    Organisation:

    State:

    Thank you all for the information and advice you have provided. It has been very helpful.

    Beth

    Beth Bint

    Clinical Nurse Consultant | Infection Management and Control Service
    Level 1 Lawson House, Wollongong Hospital 2500, NSW
    Tel. 02 4222 5869 | Fax. 02 4222 5367 | beth.bint@sesiahs.health.nsw.gov.au

    [cid:image001.jpg@01CE4293.33FF5400]

    From: ACIPC Infexion Connexion [mailto:AICALIST@AICALIST.ORG.AU] On Behalf Of Beth Bint
    Sent: Wednesday, 24 April 2013 2:05 PM
    To: AICALIST@AICALIST.ORG.AU
    Subject: Benchtop steriliser water and steam quality specifications.

    Hi All

    Hoping someone can offer some advice.

    We have not be able to find definitive specifications for water and steam quality required for benchtop sterilisers (Dental Autoclaves).

    We have two questions:

    1. Some manufacturers suggest the use of distilled or de-ionised water. If a water distiller is used how is the water quality controlled when refilling the reservoir?

    2. What are the steam quality specifications for benchtop sterilisers, and where is the reference for these?

    Thank you for any assistance you can provide.

    Beth

    Beth Bint

    Clinical Nurse Consultant | Infection Management and Control Service
    Level 1 Lawson House, Wollongong Hospital 2500, NSW
    Tel. 02 4222 5869 | Fax. 02 4222 5367 | beth.bint@sesiahs.health.nsw.gov.au

    [cid:image001.jpg@01CE40F2.33F57270]

    ———————————————————————————————

    Illawarra Shoalhaven Local Health District, South East Sydney Local Health District and Sydney Children’s Hospital Network (Randwick Campus) Confidentiality Notice

    This email, and the files transmitted with it, are confidential and intended solely for the use of the individual or entity to whom they are addressed. If you are not the intended recipient, you are not permitted to distribute or use this email or any of its attachments in any way. We also request that you advise the sender of the incorrect addressing.

    This email message has been virus-scanned. Although no computer viruses were detected, Illawarra Shoalhaven Local Health District, South East Sydney Local Health District and Sydney Children’s Hospital Network (Randwick Campus) accept no liability for any consequential damage resulting from email containing any computer viruses.

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    ———————————————————————————————

    Illawarra Shoalhaven Local Health District, South East Sydney Local Health District and Sydney Children’s Hospital Network (Randwick Campus) Confidentiality Notice

    This email, and the files transmitted with it, are confidential and intended solely for the use of the individual or entity to whom they are addressed. If you are not the intended recipient, you are not permitted to distribute or use this email or any of its attachments in any way. We also request that you advise the sender of the incorrect addressing.

    This email message has been virus-scanned. Although no computer viruses were detected, Illawarra Shoalhaven Local Health District, South East Sydney Local Health District and Sydney Children’s Hospital Network (Randwick Campus) accept no liability for any consequential damage resulting from email containing any computer viruses.

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

Viewing 5 posts - 1 through 5 (of 5 total)
  • The forum ‘Infexion Connexion’ is closed to new topics and replies.