    Dear All,

    I am seeking some assistance with interpreting the ASUM guidelines and their requirements for US probe cleaning.

    We are using the Trophon for all intracavity probes and wiping over all other probes used for routine imaging, biopsy guidance etc with a disinfectant wipe. The current wipe in use is not one of the four recommended disinfection methods referred to in their guidelines.
    We use probes covers routinely for all interventional procedures or on mucous membranes, wounds etc.

    Am I correct in interpreting the ASUM guidelines – that all probes regardless of cover usage must be cleaned in one of the four ways stated by ASUM. Can anyone throw some more light on this issue for me please.

    We have received full Imaging reaccreditation and our current practise was not raised as an issue.

    Regards

    Phillipa Parsons
    Infection Prevention and Control Clinical Coordinator
    Cabrini Health
    183 Wattletree Rd
    Malvern Vic 3144
    03 9508 1577
    0400 369 741

    Please consider the environment before you print this e-mail.
    —————————————————————————————————————
    This email and any attachments may be confidential, and are intended
    solely for the use of the individual(s) or entity to whom they are
    addressed. If you are not the intended recipient of this communication,
    please notify the sender immediately and delete the email and any
    attachments.
    Cabrini does not guarantee that this email is virus or error free.
    ______________________________________________________________________
    This email has been scanned by the Symantec Email Security.cloud service.
    For more information please visit http://www.symanteccloud.com
    ______________________________________________________________________

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    #70020
    Diane Hobday
    Participant

    Author:
    Diane Hobday

    Email:
    Diane.Hobday@NATA.COM.AU

    Organisation:

    State:

    Hello

    All intracavity probes *must* be disinfected using a high level disinfectant included on the ARTG as such – in essence, one of the four methods listed in the ASUM guidelines.

    This disinfection is in addition to the use of covers. Covers should never be considered an alternative to appropriate cleaning and disinfection.

    For external intact skin a less stringent process, in addition to a cover, can be used because these do not require ‘high level’ disinfection. (see TGO 54 for more information about what does and does not require high level disinfection)

    I hope this helps.

    Cheers
    Diane

    Diane Hobday

    Accreditation Manager – DIAS
    National Association of Testing Authorities, Australia
    1/675 Victoria Street
    Abbotsford VIC 3067

    03 9274 8200
    http://www.nata.com.au

    Dear All,

    I am seeking some assistance with interpreting the ASUM guidelines and their requirements for US probe cleaning.

    We are using the Trophon for all intracavity probes and wiping over all other probes used for routine imaging, biopsy guidance etc with a disinfectant wipe. The current wipe in use is not one of the four recommended disinfection methods referred to in their guidelines.
    We use probes covers routinely for all interventional procedures or on mucous membranes, wounds etc.

    Am I correct in interpreting the ASUM guidelines – that all probes regardless of cover usage must be cleaned in one of the four ways stated by ASUM. Can anyone throw some more light on this issue for me please.

    We have received full Imaging reaccreditation and our current practise was not raised as an issue.

    Regards

    Phillipa Parsons
    Infection Prevention and Control Clinical Coordinator
    Cabrini Health
    183 Wattletree Rd
    Malvern Vic 3144
    03 9508 1577
    0400 369 741

    Please consider the environment before you print this e-mail.
    —————————————————————————————————————
    This email and any attachments may be confidential, and are intended
    solely for the use of the individual(s) or entity to whom they are
    addressed. If you are not the intended recipient of this communication,
    please notify the sender immediately and delete the email and any
    attachments.
    Cabrini does not guarantee that this email is virus or error free.
    ______________________________________________________________________
    This email has been scanned by the Symantec Email Security.cloud service.
    For more information please visit http://www.symanteccloud.com
    ______________________________________________________________________

    ______________________________________________________________________
    NATA IT Dept e-mail system: Scanned for virus and spam (using MessageLabs Security System).
    ______________________________________________________________________
    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    _____________________________________________________________________
    This e-mail (and any files transmitted with it) is confidential and intended only for the person or entity to which it is addressed. If you have received this email and you are not the intended recipient, you must not use, disclose, distribute, copy, print or otherwise disseminate this information. Please inform the sender immediately by email and then delete this message from your system.
    Any views or opinions presented in this email are solely those of the author and do not necessarily represent those of NATA.

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    #70019
    Scott Pabst
    Participant

    Author:
    Scott Pabst

    Email:
    scott@ASHMED.COM.AU

    Organisation:

    State:

    Dear Phillipa,

    Please find links to:

    1) 2012 ASUM statement regarding the use of the wipes you refer to

    2) Current Policy position of Australian Sonographers Association on the use of the wipes you refer to

    3) WA Health operational directive regarding the re-processing of Ultrasound Transduces (including specific reference to the wipes you refer to).

    http://www.asum.com.au/newsite/Files/Documents/Policies/2012.05.28_SOP%20Press%20Release%20Tristel.pdf

    http://www.a-s-a.com.au/fileRepository/files/Website/Public%20site/Quality%20Practice/ASA%20Guidelines/0295_ASA-Guideline_ID_Update%20May2012.pdf

    http://www.health.wa.gov.au/circularsnew/pdfs/12913.pdf

    Yours Sincerely

    Scott Pabst
    National Sales Manager
    AshMed Pty Ltd
    Ph. 0435 843 950
    E. scott@ashmed.com.au

    Dear All,

    I am seeking some assistance with interpreting the ASUM guidelines and their requirements for US probe cleaning.

    We are using the Trophon for all intracavity probes and wiping over all other probes used for routine imaging, biopsy guidance etc with a disinfectant wipe. The current wipe in use is not one of the four recommended disinfection methods referred to in their guidelines.
    We use probes covers routinely for all interventional procedures or on mucous membranes, wounds etc.

    Am I correct in interpreting the ASUM guidelines – that all probes regardless of cover usage must be cleaned in one of the four ways stated by ASUM. Can anyone throw some more light on this issue for me please.

    We have received full Imaging reaccreditation and our current practise was not raised as an issue.

    Regards

    Phillipa Parsons
    Infection Prevention and Control Clinical Coordinator
    Cabrini Health
    183 Wattletree Rd
    Malvern Vic 3144
    03 9508 1577
    0400 369 741

    Please consider the environment before you print this e-mail.
    —————————————————————————————————————
    This email and any attachments may be confidential, and are intended
    solely for the use of the individual(s) or entity to whom they are
    addressed. If you are not the intended recipient of this communication,
    please notify the sender immediately and delete the email and any
    attachments.
    Cabrini does not guarantee that this email is virus or error free.
    ______________________________________________________________________
    This email has been scanned by the Symantec Email Security.cloud service.
    For more information please visit http://www.symanteccloud.com
    ______________________________________________________________________
    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    #70021
    Roel Castillo
    Participant

    Author:
    Roel Castillo

    Email:
    Roel.Castillo@LIFEHOUSERPA.ORG.AU

    Organisation:

    State:

    Hi Philipa

    With all due respect to ASUM, reprocessing of reusable medical and surgical devices and equipment must also comply with AS 4187:2003.
    As long as you satisfy the reprocessing requirements for the devices intended use as per Spauldings classification, this process was validated, to ensure the level of cleanliness, degree of disinfection and the sterility assurance level of 10( at least). As far as I know ultrasound probes fall under either the category of semi critical items (being in contact with non-sterile tissue and mucous membranes) shall undergo disinfection (thermal is preferred if available) or non-critical items (being in contact with intact skin) requiring cleaning (mechanical preferred over manual). Please identify intended use and remember this are minimum requirements for reprocessing. Some may need sterilisation, in this regard if you have an SSD just let them reprocess you instruments, their washers doesnt only clean but provide thermal disinfection as well. You have options to sterilise them as best practice entails.
    How do you validate your wiping process? Do you do random visual checks/residue checks? Are they being recorded? Do you have a procedure that this process is repeatable every time with the same level of cleanliness?
    You can do random tests, that is every other third processed instrument of the third day (or any random design) and do protein/blood/carbohydrate residue checks and document them. What measures do you take in case these checks fail?

    For better patient care
    Roel

    Roel Castillo
    Project Officer – SSD
    Mob: +61 434 496 829

    From: ACIPC Infexion Connexion [mailto:AICALIST@AICALIST.ORG.AU] On Behalf Of Parsons, Phillipa
    Sent: Tuesday, 21 May 2013 11:17 AM
    To: AICALIST@AICALIST.ORG.AU
    Subject: ASUM guidelines and US probe cleaning

    Dear All,

    I am seeking some assistance with interpreting the ASUM guidelines and their requirements for US probe cleaning.

    We are using the Trophon for all intracavity probes and wiping over all other probes used for routine imaging, biopsy guidance etc with a disinfectant wipe. The current wipe in use is not one of the four recommended disinfection methods referred to in their guidelines.
    We use probes covers routinely for all interventional procedures or on mucous membranes, wounds etc.

    Am I correct in interpreting the ASUM guidelines – that all probes regardless of cover usage must be cleaned in one of the four ways stated by ASUM. Can anyone throw some more light on this issue for me please.

    We have received full Imaging reaccreditation and our current practise was not raised as an issue.

    Regards

    Phillipa Parsons
    Infection Prevention and Control Clinical Coordinator
    Cabrini Health
    183 Wattletree Rd
    Malvern Vic 3144
    03 9508 1577
    0400 369 741
    Email: pparsons@cabrini.com.au

    Please consider the environment before you print this e-mail.
    —————————————————————————————————————
    This email and any attachments may be confidential, and are intended
    solely for the use of the individual(s) or entity to whom they are
    addressed. If you are not the intended recipient of this communication,
    please notify the sender immediately and delete the email and any
    attachments.
    Cabrini does not guarantee that this email is virus or error free.
    ______________________________________________________________________
    This email has been scanned by the Symantec Email Security.cloud service.
    For more information please visit http://www.symanteccloud.com
    ______________________________________________________________________

    ______________________________________________________________________
    This email has been scanned by the Symantec Email Security.cloud service.
    For more information please visit http://www.symanteccloud.com
    ______________________________________________________________________
    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

    ______________________________________________________________________
    This email has been scanned by the Symantec Email Security.cloud service.
    For more information please visit http://www.symanteccloud.com
    ______________________________________________________________________

    Messages posted to this list are solely the opinion of the authors, and do not represent the opinion of ACIPC.

    Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

    Replies to this message will be directed back to the list. To create a new message send an email to aicalist@aicalist.org.au

    To send a message to the list administrator send an email to aicalist-request@aicalist.org.au.

    You can unsubscribe from this list be sending ‘signoff aicalist’ (without the quotes) to listserv@aicalist.org.au

Viewing 4 posts - 1 through 4 (of 4 total)
  • The forum ‘Infexion Connexion’ is closed to new topics and replies.