Hi Kylie,

Hollow Type A and B are still design considerations for small steam sterilisers and are relevant while AS4815 is still in play. The design and requirements of these sterilisers are covered by ISO13060, which will continue to apply as a normative standard, even if the overarching reprocessing standard in Australia were to change. With a beaver handle, the blind ended hollow is similar to a number of other challenging little nooks and crannies. There are plenty of RMDs that may not even conform to the ideal picture of lumens or hollows but have (either hidden or obvious) close mating or narrow gaps and spaces.

ISO13060 contains the definitions and diagrams of simple hollow items (from which Type A and B are derived) and how to assign them based on the ratio of length to diameter of the cavity. This was inserted verbatim into AS4815.
Annex A of ISO13060 has a further graph to illustrate the way RMD structures are defined for the purposes of testing and classifying the load contents of small steam sterilisers, which is interesting and worth a look. The axes are the diameter and length of the RMD’s feature, and lines representing the length:diameter ratios that define Type A and B Hollow items are used to describe three plot areas into which you could place your RMD – these being non-hollow, simple hollow and narrow lumen areas. The l:d of double ended and single ended lumens are also differentiated in this. I haven’t measured up the screw hole of a beaver handle but it looks like it would come under “narrow lumen” field on that chart.

S Type sterilisers include cycles for which the manufacturer has specified the load type and limitations, so those cycles have to be followed according to the manufacturer’s instructions, including testing requirements. That aside, if I were considering the beaver handle as having a blind, narrow lumen then air removal testing would be warranted.

Regards,

Andrew Ellis
Sterilising and Reusable Medical Device Reprocessing State Coordinator
Infection Control Service | Communicable Disease Control Branch
Health Regulation & Protection
Department for Health and Wellbeing | Government of South Australia
Level 13 | 25 Grenfell Street | Adelaide SA 5000
HCW infection prevention: http://www.sahealth.sa.gov.au/infectionprevention
General public: http://www.sahealth.sa.gov.au/hospitalinfections

This email may contain confidential information, which also may be legally privileged. Only the intended recipient(s) may access, use, distribute or copy this email. If this email is received in error, please inform the sender by return email and delete the original. If there are doubts about the validity of this message, please contact the sender by telephone. It is the recipient’s responsibility to check the email and any attached files for viruses.

Hi everyone,

I wanted to ask the group about reprocessing a specific podiatry instrument.

The instrument is a beaver chuck blade/handle. This item is hollow but capped at the end and when in use the blade holder is screwed into the handle and the blade attached.
It’s getting reprocessed (handle separated from the blade holder) in a S Type benchtop steriliser in a AS/NZS4815 environment.

See attached image of the instrument.

Anyway, would this item be considered as a hollow type A or type B and is that even a thing anymore?

Essentially, I am wondering about having to conduct air removal steam penetration tests (ARSPT) for these S Type benchtop sterilisers because of these items. Looking at 4815 it says that if it’s a hollow A or B item then the steriliser needs an air removal steam penetration test anyway. I guess my question is would the ARSPT for this S Type steriliser each day be required because the item is a Type A Hollow or Type B Hollow and why. I’d like to know the rationale for decision making I suppose.

I have been reflecting on this and I think that because the item is hollow, it doesn’t make a difference whether it’s hollow A or B because it’s, well, hollow – and the Standard says if the item is hollow (A or B) then it needs an ARSPT. I just want some more insights about the A and B hollow situation for my own interest.

Thanks for considering this one. I would welcome any offline comments and, of course, through the forum.

Kylie

Kylie Robb
MHSM, GradCert InfecPrev&Cont, CICP-A, MAICD, FACIPC
ADA NSW Head of Practice Services
Advanced Credentialed Infection Control Professional (CICP-A)
Adjunct Lecturer – University of Newcastle – Oral Health School of Health Sciences
ACIPC Board Director

Australian Dental Association NSW Branch
1 Atchison Street, St Leonards NSW 2065
t: 02 8436 9936 m: 0438628664
E: kylie.robb@adansw.com.au | W: https://www.adansw.com.au

[cid:image002.png@01D88D2F.F378F700] [cid:image003.png@01D88D2F.F378F700] [cid:image004.png@01D88D2F.F378F700]

ADA NSW acknowledge the Traditional Custodians of the land on which we work; the Cammeraygal people of the Eora Nation, and pay respects to Elders past, present and emerging.

MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au

To send a message to the list administrator send an email to admin@acipc.org.au

You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au

MESSAGES POSTED TO THIS LIST ARE SOLELY THE OPINION OF THE AUTHOR, AND DO NOT REPRESENT THE OPINION OF ACIPC.

The use of trade/product/commercial brand names through the list is discouraged by ACIPC. If you wish to discuss specific reference to products or services by brand or commercial names, please do this outside the list.

Archive of all messages are available at http://aicalist.org.au/archives – registration and login required.

Replies to this message will be directed back to the list. To create a new message send an email to acipclist@acipc.org.au

To send a message to the list administrator send an email to admin@acipc.org.au

You can unsubscribe manually from this list by sending ‘signoff acipclist’ (without the quotes) to listserv@aicalist.org.au